Hi
Please note. The original attachment was incorrect. I have added the correct block diagram.
I know this is a long post but I really need help on this. I tried to be thorough in my description.
I need some help with a faulty circuit. See the attached schematic block diagram.
With the original setup (all secondaries connected ) the 250mA rated fuse F101 blows every time that I power the unit up with S101. The LEDs on the front display come on momentarily. The display circuits are powered by the 5v display & 5v logic rails.
I then proceeded to disconnect the secondaries one by one to see when the fuse stops blowing. I started with 1400V and then 140V. During these two disconnection and testing procedures, the fuse blew just as it did with all the secondaries connected.
I then disconnected the secondary at the (5V display & 5V logic) section. After switching on I could hear the transformer hum for the first time. So the fuse did not blow. I left the unit switched on for hours before returning to it. It was still humming away.
I then proceeded with fault finding and this is where I lost the testing plot a bit. I assumed that it must be the circuit connected to this (5V display & 5V logic) secondary that caused the problems. So during my connecting and disconnecting the secondary here while removing and testing components
I never checked that the unit can be switched on multiple times when the secondary is disconnected. It turned out that the fuse blows even with this secondary disconnected. Later it turned out you just need to cycle the power multiple times for the fuse to blow.
So after the confusion of this last test, I decided to disconnect all secondaries. I.e. the last one to be removed was the (+/- 15V and 24V).
Now with all primaries disconnected and a new fuse, the unit switched on. I left it on for a while and switched it off for 3 seconds. Then switched it on again and the fuse blew. I repeated this test with multiple 250mA fuses. The only variation was that sometimes I had to cycle the power more than once to blow the fuse. But not in one case did a fuse blow on the initial switch-on.
I also tried a 500mA fuse as well. This one blew on the first attempt.
The 250mA fuses are from old stock. Brand = Fusit.
The 500mA fuse is new. Brand = OneHungLow
Their all fast-blow fuses.
I then checked the transformer with a Agilent U1733C
Primary: L=4.4 Henry R = 27 Ohm As measured on the live and neutral input pins of the line filter. These seem like decent values.
Sec 1400V: L = 4.1 Henry R=2.18k Ohm
Sec 140V: L = 3.9 Henry R= 81 Ohm
Sec 24v&15V: L = 336 milliHenry R = 3.7 Ohm
Sec 5V: L = 27milliHenry R = 0.4 Ohm
All inductances measured at 100Hz.
The Sec 5V: R = 0.4 Ohm seems a bit low?
I also measured the insulation between all primary to chassis ground and all secondaries to chassis ground. There are no shorts. I don't think I need to measure this with a megger since 250mA current to chassis would have triggered my earth leakage breaker.
I have not tested the line filter. Part number = 3EF1 by Systron donner.
It looks like this part.
https://www.ebay.com/itm/133555373217I assume it is this part.
https://www.mouser.co.za/ProductDetail/TE-Connectivity-Corcom/3EF1?qs=sfrgH6goU2Jyc6W8Z1nXxw%3D%3DI assume the filter should be fine since the fuse sits after the line filter. So any short between live and neutral in the filter will not be seen by the fuse.
Also, I have measured the primary resistance and inductance (as shown above) through the live and neutral pins of the line filter and the values seemed fine. No shorts. I have also measured the capacitance on the live and neutral pins = 660nF @ 100Hz.
What could be the problem here?
Does anyone have ideas on where I should look next?
The complete user manual for the instrument can be found here.