'Production variations in emitter-base voltage will typically cause about ±10°C variation. Since the emitter-base voltage changes about 2mV/°C and is very predictable, other temperatures are easily set.'
Has anyone measured the actual operating temperature of their LTZ1000(A)? I guess that might not be easy, especially with the LTZ1000A - would it be better to measure the collector current against temperature when heating the device externally?
I'm asking because for long term stability it is best to run the device at the lowest temperature that can be maintained at the maximum ambient so 10C of variation is quite significant. The LTZ1000 part appears to have the advantage here as well as being cheaper. Apart from the higher power consumption are there other disadvantages - such as bigger thermal problems due to higher lead temperatures?
Thanks, Splin
Well, the dU/dT of UBE is very predictable, i.e. -2mV/°C.
The variation of the absolute value of UBE is what causes the 10°C variation.
Therefore, I have assembled the LT circuit, and first disabled the heater part.
Then I attached a fast DMM, with MiniMax function, the 34401A, to measure the UBE of the voltage reference transistor, that is Q1.
I noted the ambient temperature, that's the same, the circuit is initially residing on.
Powering up, the DMM measures the UBE, let's say 0.560V, which will rapidly decrease by self heating of the reference circuit.
These 0.56V are stored as a maximum value, representing the measured room temperature.
You can repeat that measurement several times for higher reliability of the R.T. value.
If you now engage the heater (12k/1k for 45°C in my case), the UBE will further decrease with relatively precise -2mV/K, and you now can easily calculate the real stabilization temperature.
I do not remember the outcome of this measurement, but the calculated setpoint was quite precise.
It's also not that dramatic to have 5..10°C more.
Both of my LTZ1000s drift less than -1ppm/yr with blindly assembling 12k/1K, and only big variations as 65°C or these 95°C of the 3458A will noteworthy degrade the performance.
Therefore, I don't recommend to trim the oven temperature exactly to 45°C.. the exemplar variation also causes different stability figures.
Only make sure, that the temperature is not less than 45°C, because of regulation stability.
And more important is, to avoid big temperature excursions, due to soldering, or due to wrong oven temp, or due to oven regulation faults, because of the big hysteresis of the device.
Frank