I need to find a non-battery power solution for a very sensitive scientific machine. The machine uses two compressors and is designed to use two single phase Australian 230 VAC, 50Hz inputs - one input for each compressor.
However, this machine is going to be demonstrated in the USA where the nominal mains voltage is a single phase 110VAC, 60Hz. A step-up transformer is no good because it would change the frequency. I have been told (to be confirmed) a modified sine wave output is no good either, but a pure sine wave output is required. I do not know what, if any, feedback mechanisms are in the equipment, but if it were used in Australia, typical variances in mains amplitude obviously isn't an issue.
There are two 230V compressors, one draws 6A and the other 4A. I have also been told "the inrush current is 20A for as long as 200ms". I assume that is for each mains input - more to the 6A compressor I assume.
The machine can have its code modified so only one compressor switches on at a time, hence the inrush current at any one time need not exceed 20A.
I want to use something off-the-shelf, but if required we can build such a converter in a box with two 230V mains outlets and one (maybe two) 110V mains inlets and get it electrically certified. But maybe something is already complete ready to work. Of course, two independent power supply converters would work too, but one might be a more attractive option.
This device needs to be reliable. I searched in the WWW and it is flooded with solar inverters made by every man and his dog.
I normally work in microelectronics and PCB design, but not much in this area. If anyone has any suggestions or tips on where to get such a mains power converter, please let me know.
cheers,
Dave
First of all, the US mains are split phase 240 VAC, 60 Hz. Since the 60 Hz, will probably make the motors on the compressors run faster, your ramp times, etc. will be different. Most wall units in the office or hospital rooms will only have 120 VAC, 60 Hz. outlets, however, if you are doing your trials in a lab like room, it may have the 240 VAC 60 Hz. outlets or can easily be installed.
Second, you said that there are 2, 230 VAC, 50 Hz, inputs, one for each compressor which means that your solution, whether it is a VFD or something else will require 2 separate outputs. Most 240 VAC outlets in the US are single outlets. If you can use a single 240 VAC, 60 Hz. input on your VFD or other solution, no worries. There are different current ratings that are common for 240 VAC outlets: 30 amps and 50 amps, which each should have the proper sized wire to accommodate these currents.
Third, If you come up with a 120 VAC, 60 Hz. solution, keep in mind that most 120 VAC wall outlets in the US are rated at 15 amps. There are 20 amp outlets available, but the NEMA configuration for the power chord is not common and would be a "special" installation, which may be a deal breaker.
Fourth, if you know where you will install this unit, you might talk to the maintenance or technical liason in the building and this person may work a solution for you. It can't hurt to ask. Most places in the US are very accommodating and understand the 60 Hz/50 Hz dilemma.
I know this didn't answer your original question, but at least you have more facts about US power outlets, etc.
Hope this helps...