ejeffrey:
RC timing is fine if you don't need much accuracy and only have digital pins and comparators to work with.  It isn't a good way for accurate resistance measurement (OP specified 0.1%).  You are using a capacitor as a reference instead of a resistor and capacitors generally have orders of magnitude worse accuracy, drift, and linearity than even inexpensive resistors.  Furthermore, you can't do a 4-terminal measurement and the GPIO pin on resistance is in series with the DUT and can generate a large error.

For exactly these reasons RC time is a good and common way to measure capacitors.  Due to their generally poor tolerances it is rarely needed to measure capacitors much better than 1% and resistors of that accuracy are easily available.
babysitter:

--- Quote from: babysitter on September 16, 2019, 05:58:23 pm ---(Additionally, discharge/charge thru a known good capacitor sometimes for reference.)

--- End quote ---

This should have been resistor.
Kleinstein:
The measurement of the resistance via a delay gets tricky due to the switch resistance. So it may be practical for large values, but not for small ones like 1 K or less. In addition at high resistance parasitic capacitance can change things.
mzzj:
Layout? external wires? thermovoltages?

Tie 3.3v to ADC ref input and 3.3k precision resistor, 1mA to 10 ohms  gives you 10uV per 0.1%, easy peasy for something like LTC2484. Upper end of the range is easy after that.
splin:

--- Quote from: mzzj on September 17, 2019, 04:40:26 pm ---Layout? external wires? thermovoltages?

Tie 3.3v to ADC ref input and 3.3k precision resistor, 1mA to 10 ohms  gives you 10uV per 0.1%, easy peasy for something like LTC2484. Upper end of the range is easy after that.

--- End quote ---

Better still, use a 447ohm precision resistor to give 70.6uV (21.4ppm) per 0.1% change in sensor resistance at both 10 ohms and 20k, and 500+uV/0.1% from 100 to 2000 ohms with the best sensitivity at 447 ohms. Change the reference to get the best sensitivity to match your application needs.

A higher value reference resistor is likely to be preffered given that:

a) Most ADCs have different errors at zero compared to full scale. Eg, the LTC2484 has 2.5uV offset error but 82.5uV max gain error before calibration + .33uV/C gain drift (with 3.3V reference).