I recently bought a 3437A System Voltmeter at a bargain price.
The very first step was to remove the covers and look for the Rifa firecrackers. None!
After switching it on, all the buttons lit up and all the segments on both displays came on, but there was no response to any of the knobs.
Turning it on was a mistake, though. It’s unclear how long the previous owner had been running the device in this state – so, effectively, I had bought the damage along with it.
What bothered me was that HP had, without exception, mounted all ICs in the same orientation relative to pin 1 – but apparently not the main processor (a custom part, the white-ceramic version).
So I took screenshots from a few YouTube repair videos, and it became clear: the processor had been fitted the wrong way round in its socket.

Normally, that’s an instant kill!
Once fitted correctly, not much changed, apart from a single “E” appearing on the display.
I then continued searching for the actual fault – and found and repaired it. The two Sprague 220 µF capacitors supplying the –6 V (and –3 V for the processor) were completely dried out. I measured just 160 nF + 50 pF.
The display was then very nervous, but all the knobs were working. However, the displayed voltage was approximately 1.5 times too high.
Cleaning the very dirty acquisition board and replacing all the electrolytics in it finally resulted in a stable zero reading, and the displayed voltages are now close to a 289.
It’s not really worth talking about a routine repair, but all kudos to HP!

I never thought a processor powered backwards would survive – but it did!

(image shows processor mounted in right orientation!)