Thank you!
So, I will not worry about the resistance measurement, mostly because without excitation applied I read 350ohm +-2ohm between ranges, that is ok to me.
Also, thank you for making me notice the zero position drift due to layout.
What sounds strange to me is that:
- the absolute error is almost the same between rages, and it's twice the mV absolute error in range 0.0
- the absolute error is linear and depends on the excitation voltage
This seems something not fully related to component quality/accuracy, but to the design of the device. Am I wrong saying this?
Excitation 10.024V readings (expected, read and abserr in mV)
range expected read abserr
0.0 0.000 0.181 0.181
0.5 5.012 5.311 0.299
1.0 10.024 10.316 0.292
1.5 15.036 15.314 0.278
2.0 20.048 20.320 0.272
Excitation 4.999V readings (expected, read and abserr in mV)
range expected read abserr
0.0 0.000 0.087 0.087
0.5 2.500 2.647 0.148
1.0 4.999 5.146 0.147
1.5 7.499 7.643 0.145
2.0 9.998 10.142 0.144