Overspeed:
Hello
Thanks for explanation I agree but
a strain gage is a not a rubber band , a strain gage have a gage factor so resistance variation is limited and limited also by the fixation / gluing so a simulator will match the reality no need of a wide resistance variation but better to have an very accurate link between Delta resistance and output voltage of the bridge
AC bridge request ( or have in the past as HBM AC bridge ) complex electronic with some advantages but I use DC since years that work easy
I link a interesting doc on 1/4 bridge and a symmetrical dual op amp circuit
I agree on the output voltage to voltage ref ratio solution
Regards
OS
hozone:
Thanks again.
The DAC, or any other active component would work better for this design.
I think I'll go for a +-0.003% VExc absolute error label for this passive simulator. Even if it's a unknown behavior to me :palm:
Overspeed:
--- Quote from: hozone on August 06, 2024, 07:54:27 am ---Thanks again.
The DAC, or any other active component would work better for this design.
I think I'll go for a +-0.003% VExc absolute error label for this passive simulator. Even if it's a unknown behavior to me :palm:
--- End quote ---
Hello
Can you provide the calculation : I'll go for a +-0.003% VExc absolute error label
That interesting
Regards
OS
hozone:
Hi @Overspeed,
Don't know if I'm right, and why it's like so, but:
Excitation 10.024V readings (expected, read and abserr in mV)
range expected read abserr
0.0 0.000 0.181 0.181
0.5 5.012 5.311 0.299
1.0 10.024 10.316 0.292
1.5 15.036 15.314 0.278
2.0 20.048 20.320 0.272
Excitation 4.999V readings (expected, read and abserr in mV)
range expected read abserr
0.0 0.000 0.087 0.087
0.5 2.500 2.647 0.148
1.0 4.999 5.146 0.147
1.5 7.499 7.643 0.145
2.0 9.998 10.142 0.144
Locking at the results above (for all ranges expect 0.0):
* absolute error is almost 0.146(mV) when Vexc is 5V (5000mV), so 0.00292%, let's say 0.003% * VExc => 0.15mV
* absolute error is almost 0.285(mV) when Vexc is 10V (10000mV), so 0.00285%, let's say 0.003% * VExc => 0.3mV
Note that absolute error for 0.0 range is almost 1/2 (0.003% * VExc)
hozone:
Little updated.
I've notice the 0.272mV shift even on simulator on the R1/R7/R5/R8 branch, given a Vexc of 10V. (see image).
Is it somehow correlated? :-//