| Electronics > Projects, Designs, and Technical Stuff |
| Pressure sensor device amplification |
| (1/2) > >> |
| jts1705:
Hello everybody, I am a student who is developing a pressure sensor device for a medical application. I need to measure pressures between 0 and 50mmHg. Therefore, I am using arterial pressure sensors that suit my purposes (http://www.icumed.com/products/critical-care/blood-pressure-monitoring/transpac-it.aspx). I need to amplify, filter and condition the signal before Arduino UNO can read it. I've designed an schematic that I think can work, but now I am trying everything in a protoboard and it does not work. I have tested with a voltmeter partial voltages and the DCDC converter seems to be not working good. It does not give any output signal. I have checked again its datasheet and seems like it has a CTRL pin... but I do not really think it is the reason why it does not work. (https://eu.mouser.com/datasheet/2/468/RS-S_D_Z-958418.pdf). I do not know if the DCDC converter is broken or if I have performed a mistake before... I would be very very very thankful if any of you could give me his/her opinion. I attach the schematic and two photos of the protoboard. Thank you very much!! |
| jbb:
I suggest you divide and conquer. You can gently pull out the RS1212D and test it separately. Note that the way your schematic is drawn, the RS1212D is not providing any isolation; the grounds are connected. Also something’s weird with the pin numbers - I’m pretty sure the datasheet doesn’t go up to 23 pins on it. I also have a concern about the instrumentation amplifier. I may be wrong but the high resistance on AR1 pin 5 could lead to poor results. Normally an INA needs a low resistance on the ‘ground’ connection (a few Ohms or less) to avoid disturbing the differential amp. |
| babysitter:
I don't know the icumed from the top of my mind but IBP transducers usually have a pinout where the excitation and signal lines are "intertwined" like supply+ signal- supply- signal+ or similar, but the supply together and signal together looks very wrong to me. Between the suppy in pins you should measure a higher (>1k-ish) resistance than between the output pins. When applying power supply, the signal lines should be close to 1/2 supply voltage relative to ground. Consider adding at least some capacitor between the Instrument Amplifier reference and ground, they like lower impedance here. That nice schematic in the datasheet, where the offset compensation voltage is buffered with a OpAmp, is there for a reason. Good luck! Hendrik |
| jts1705:
Thanks to all. I followed all your instructions and I considered to remake the schematic and correct its mistakes. (I do not have Altium license anymore so I did it as I could). -->I have corrected the pins in the DCDC converter according to the datasheet and also the pins in TL062 (although they were wrong in the schematic I think I was doing it right in the protoboard). I do not really know why, but now the DCDC converter seems to be working but it gives almost +14V in Out+ and -10.10V in Out-. Instead of +12V and -12V that is what I was expecting. However, I think this should be enough to supply to the INA128 and TL062. But stills not working. :( I will definitely lower the impedance in AR1 to not disturb INA128. Thanks for that!! |
| aldi:
Check section 8.2.2.4 in the INA128 datasheet, depending on the output structure of the pressure sensor, you may need to add a bias current return path. If the pressure sensor is just a wheatstone bridge this is not needed. |
| Navigation |
| Message Index |
| Next page |