Hi
a small follow-up ,got on ebay a old "minXray 210" portable x-ray generator for good price
unfortunately unit not working and was very unsafe ,brain of machine logic was a 555 in monostable ;-)
but chip was die and shoot x-ray as long you keep finger on trigger button !!! (not what to be last patient on that)
and i also found 2 broke wire in the sealed head that contain HV transformer and tube
one wire was current resistor tap to ammeter and second one was one of the transformer tap
so fix back the broke wire and the replace the 555 chip ,but still noting, unit draw lot of current but not blow fuse
made quick math and it may normal 63Kv @ 12ma = ~ 800W ,but transformer humm and vibrate allot
and Dental CCD still not trig and my old geiger counter to so i assume it not generate X-RAY ,even if i treat it as it work
next week-en will re-open sealed case and make in deep test on HV transformer for try to see what wrong
my idea was feed it @1V rms 60 HZ and make sure have 630volt at output
not have clue yet for test filament (tube was inside another tube and pain to access ..
p.s. tube seem to not draw current even if ammeter wire was fix ,i validate the panel meter and it accurate and work
in transformer it have 1K resistor in serie whit GND side and current meter plug on it , since meter internal resistance was low
i assume the 1K was here only as safety if ammeter become open , i test panel meter whit current source @10ma and it show 10ma
Best regard