Repair / Re: TDS744A: 1M input impedance measures: Ch3 = 160Kohm, Ch4 = 706KOhm
« Last post by John_ITIC on Today at 03:43:14 am »I am suspecting that soldering the relays in may have heat damaged the three 1M voltage divider resistors. This post lead me to do some more experiments:
https://www.eevblog.com/forum/testgear/tektronix-tds520b-attenuator-swap/
This lead me to try this experiment myself. When I apply some DC voltage to the 1M input (some 19V), I'm seeing that the input impedance rises several hundred Ohms. I can even go over 1MOhm. I then measure the input resistance with a couple of DMMs, one show around 1MOhm while the other show some 800 KOhm. The values drift towards etch other as I swap meters. It looks like the meter voltage is different and that the input resistive divider reacts to this change in meter current.
I'm seeing the same behavior on both Ch3 and Ch4 that I swapped out the relays on. Ch1 and Ch2 are stable across temperature and across multimeters (both show 998 KOhm input resistance). Ch1-2 do not get affected by applying a DC voltage.
So, has anyone seen similar issues where soldering iron heat has permanently affected the temperature coefficient of resistors in such a way?
Note: I'm also seeing on Ch3 and Ch4 that setting the coupling impedance to GND still measures 1.7MOhm on Ch3 and 17 MOhm on Ch4.
Both Ch1 and Ch2 measure infinite resistance when set to GND.
So, it seems there is some shunt resistance that is related after all. But where?
Thanks,
/John.
https://www.eevblog.com/forum/testgear/tektronix-tds520b-attenuator-swap/
Quote
Sometimes, input impedance of CH1 is 170ohms in 1Mohm mode and 38ohm in 50ohm mode as if a 170ohm resistor was in parallel with the input. Apart from this, attenuator behaves correctly.
The strange thing is that this fault disappears when I apply some DC voltage (12V for example): after that, the input impedance is back to normal (1Mohm or 50ohm) until powered off.
This lead me to try this experiment myself. When I apply some DC voltage to the 1M input (some 19V), I'm seeing that the input impedance rises several hundred Ohms. I can even go over 1MOhm. I then measure the input resistance with a couple of DMMs, one show around 1MOhm while the other show some 800 KOhm. The values drift towards etch other as I swap meters. It looks like the meter voltage is different and that the input resistive divider reacts to this change in meter current.
I'm seeing the same behavior on both Ch3 and Ch4 that I swapped out the relays on. Ch1 and Ch2 are stable across temperature and across multimeters (both show 998 KOhm input resistance). Ch1-2 do not get affected by applying a DC voltage.
So, has anyone seen similar issues where soldering iron heat has permanently affected the temperature coefficient of resistors in such a way?
Note: I'm also seeing on Ch3 and Ch4 that setting the coupling impedance to GND still measures 1.7MOhm on Ch3 and 17 MOhm on Ch4.
Both Ch1 and Ch2 measure infinite resistance when set to GND.
So, it seems there is some shunt resistance that is related after all. But where?
Thanks,
/John.