Repair / Re: TDS744A: 1M input impedance measures: Ch3 = 160Kohm, Ch4 = 706KOhm
« Last post by John_ITIC on Today at 07:15:21 pm »Okay, this issue has now been resolved. It turns out that the solder flux I used (Kester 2331-ZX) had been absorbed into the hybrid substrate. The substrate became conductive, which formed a resistive path in parallel with the 1M input impedance. I found that submerging the hybrids in IPA and scrubbing with a brush did not reverse this unwanted conductivity. The only thing that worked was to soak the hybrids in flux remover, scrub, rinse in boiling hot water, scrub some more and blow dry.
This is the flux I'm using:
https://www.kester.com/DesktopModules/Bring2mind/DMX/API/Entries/Download?Command=Core_Download&EntryId=3679&language=en-US&PortalId=0&TabId=96
This is an "organic acid, water soluble" flux and it is stated that hot water should be used for cleaning. There is no mention of flux remover or IPA. I'm not sure what the ideal cleaning procedure is.
My recommendation is to measure the resistance between the Hybrid's BNC input pad and neighboring GND tabs before soldering back into the A15 attenuator assembly. The input relays are by default in the GND mode (input floating) so there should be infinite resistance. If you have any resistance, even tens of MOhms, it means you still have contamination on the board. The black relay is responsible for the DC/AC/GND setting and it is set to GND when unpowered.
My scope now passes SPC and there is roughly 0.250mV offset between the channels when in 1mV/div setting in 1M input impedance setting. I'm not sure how to actually use the 5mV/div, 2mV/div and 1mV/div setting because the scope will not allow me to select those fine vertical resolutions when I'm plugging passive probes in (min vertical resolution then becomes 10mV/div).
So, I consider this repair successful. Hopefully it will help someone else doing the same repair.
Thanks,
/John.
This is the flux I'm using:
https://www.kester.com/DesktopModules/Bring2mind/DMX/API/Entries/Download?Command=Core_Download&EntryId=3679&language=en-US&PortalId=0&TabId=96
This is an "organic acid, water soluble" flux and it is stated that hot water should be used for cleaning. There is no mention of flux remover or IPA. I'm not sure what the ideal cleaning procedure is.
My recommendation is to measure the resistance between the Hybrid's BNC input pad and neighboring GND tabs before soldering back into the A15 attenuator assembly. The input relays are by default in the GND mode (input floating) so there should be infinite resistance. If you have any resistance, even tens of MOhms, it means you still have contamination on the board. The black relay is responsible for the DC/AC/GND setting and it is set to GND when unpowered.
My scope now passes SPC and there is roughly 0.250mV offset between the channels when in 1mV/div setting in 1M input impedance setting. I'm not sure how to actually use the 5mV/div, 2mV/div and 1mV/div setting because the scope will not allow me to select those fine vertical resolutions when I'm plugging passive probes in (min vertical resolution then becomes 10mV/div).
So, I consider this repair successful. Hopefully it will help someone else doing the same repair.
Thanks,
/John.