But can it really be as effective as reported? On their website it says:
"The team used the FREO2/SIPHON system to produce medical grade oxygen concentrations above 90%, sustaining the production of concentrated oxygen into the night."
Absolutely. As long as they have a continuous flow of water they can produce a relatively steady vacuum. The vacuum drives a bellows based "pump" that generates enough pressure to run the PSA process. Remember, they are talking air at 90% oxygen content at 1/317 the flow rate of the water used to power the process. It's not fast by any stretch. I can buy an over the shoulder battery powered generator that makes that much oxygen at 90% purity.
You don't need a massive airflow to reach 90% purity. I got close to that on my patio table with some hacked together PVC fittings and a couple of E-bay solenoid valves. You need considerably more air for each % you try and get closer to pure however (and better zeolite).
They state the test system had 540Lpm of water flow, or 9L/s. That's a pretty good flow. Oxygen at 90% purity at 1.7Lpm requires about 4.2Lpm of air at *best*. Lets be generous and figure triple that, so 12.6Lpm or 210ml/s.
They are talking 5PSI (~4M head) and that's doable with the right zeolite (the stuff I have needs ~30PSI, but it was cheap).
So they are injecting .210L/s of air into a 9L/s water flow to get the vacuum. It's not unachievable. The best part is it scales. Got more water, put more siphons in and multiply your flow.
All in all, it's a pretty cool concept. Don't forget the vacuum doesn't have to be producing pressure on a 1:1 basis. The mechanical advantage can be used to optimise the delivery pressure for the PSA cycle at the expense of volume or vice versa.