Jim from Chicago:
For the Texas Instruments TPS798xx-Q1 LDO, it is nominally rated to source 50mA. But in section 6.5 of the datasheet at the bottom it says it can source between 60mA to 200mA for ΔVOUT = VOUT(NOM) – 0.1 V (I'm using adjustable version). So is the datasheet saying that if you want to source between 60mA and 200mA, you can do it, but your Vout will droop by 100mV?
ledtester:
Link to the datasheet:
https://www.ti.com/lit/ds/symlink/tps798-q1.pdf
The 50mA figure is what the device is guaranteed to give you continuously.
I think the I_limit figure at the bottom of page 5 tells you that current limiting will be engaged somewhere between 60 mA and 200 mA.
The test condition "\$\Delta V_{OUT} = V_{OUT(NOM)}-0.1 V\$" suggests that the output voltage during the test was 0.1 V -- i.e. a low enough resistance was put on the output so that current limiting reduced the output voltage to 0.1 V.
I agree that footnote (9) is confusing in this regard.
Jim from Chicago:
So it's saying that somewhere between 60mA and 200mA, the LDO will refuse to source more current?
srb1954:
--- Quote from: ledtester on January 31, 2023, 11:39:51 pm ---Link to the datasheet:
https://www.ti.com/lit/ds/symlink/tps798-q1.pdf
The 50mA figure is what the device is guaranteed to give you continuously.
I think the I_limit figure at the bottom of page 5 tells you that current limiting will be engaged somewhere between 60 mA and 200 mA.
The test condition "\$\Delta V_{OUT} = V_{OUT(NOM)}-0.1 V\$" suggests that the output voltage during the test was 0.1 V -- i.e. a low enough resistance was put on the output so that current limiting reduced the output voltage to 0.1 V.
I agree that footnote (9) is confusing in this regard.
--- End quote ---
I would interpret that specification as saying the criteria for being in current limit is when the output voltage drops 0.1V from its nominal value, not when the output drops all the way to 0.1V. In effect, they are defining the onset of current limiting to be at that point where the output goes out to regulation by at least 100mV.
When you reread the footnote I think it makes more sense using this definition although they have used a rather clumsy way of stating it. In particular, they have not clearly defined Vo(nom). Is it the design centre voltage of the regulator at its nominal 50mA output current, or is it the actual unloaded voltage taking into account production tolerances etc, or something in between?
Siwastaja:
Current limiting is a protection feature. 50mA means maximum current you can pull with the regulator working normally and maintaining regulated output voltage.
Current limiting mode is not normal operation and by definition will drop the output voltage. You want to keep the highest load current below the minimum possible current limit onset value so that current limit never triggers.
Additionally, you need to calculate the maximum power dissipation and the resulting die temperature to see if it flies. Usually, you can only pull the rated maximum current out of linear regulators when the difference Vin - Vout is small.
P = (Vin-Vout)*Iout
Tj = Ta + RthJ-A * P