after looking on measurement place the generator Agilent E8267C is power down and no action.
Device runs for long time (45800h).
Note!! is device in standby, everytime warming the clock and can be damage by power lost or/and to fast transport!
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905338-0](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905340;image)
first check service manual 1. See no fuse external on power supply.
-> check inside , two fuse (blue) to check and the typical ceramic thermal fuse (red) are ok.
-> but design mistake, Fan blow outside (in picture below) and the warm air flow about caps, caps looks little defect (166µF on LCF)-> chance in one was inhouse.
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905342-1](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905344;image)
ok power up! but error in selftest 10MHZ
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905312-2](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905318;image)
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905316-3](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905320;image)
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905322-4](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905324;image)
-> check voltage, but all good. ok PS-fault red led on -> hm
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905326-5](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905328;image)
-> change big powersupply caps to mouser 871-B43547A9187M000, TDK long life, old was 3000h new are 8000h
871-B43547A9187M000-> so look clock board, 10MHZ clock (red marked) and on side warming resistor. Clock and Resistors are bad soldering and i fixed.
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905346-6](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905348;image)
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905350-7](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905352;image)
but after rebuild and test, the powersupply fan doesnt work, think doesnt work bevor and wasmain problem to fault device.
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905356-8](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905358;image)
changed in Conrad.biz ordernr. 1926405-VQ, EBM Papst 3412NHH with little more power.
https://www.conrad.de/de/p/ebm-papst-3412nhh-axialluefter-12-v-100-m-h-l-x-b-x-h-92-x-92-x-25-4-mm-1926405.htmlafter that looks nice.
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905360-9](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905362;image)
![[fixed] repair Agilent E8267C, power down, 10Mhz error 905364-10](https://www.eevblog.com/forum/index.php?action=dlattach;topic=225110.0;attach=905366;image)