Author Topic: [in repair now] repair HP 8114A High Power Pulse Generator 100V/2A  (Read 1212 times)
Offline tagchenTopic starter
hi, actually, or more an long time repair :),  i repair an HP 8114A.
problem, selftest  "no trigger signal for output board" , output relais doesnt work (generator say its on, but is off).
i use the generator for extremly high power (...-120KW) LIDAR prototypes and the problem is you don't get another generator for money either, mostly low power.
909776-0
909756-1
i measure the procedure from service manual with sperated transformer and scope but the big white transistor and GND resistor blow up.
change transistor, but i have no idea about resistor (test 1K), sme problem with selftest.
after that, i pick out the voltage boards and measure testpoints, -17V and 12V are mistake. after that checked power supply pcb outside and all voltage are ok
--> must problem on output board.
909760-2
909764-3
909768-4
909772-5
i will spend some money for schematic and check again. acually no ideas that i can do next. and other projects.
« Last Edit: January 17, 2020, 09:36:22 am by tagchen »
Offline Mike112
Re: [in repair now] repair HP 8114A High Power Pulse Generator 100V/2A
« Reply #1 on: February 04, 2020, 04:35:48 am »
Did you find a schematic? I'm working on fixing the same unit with the same problem right now and am looking for a service manual
« Reply #2 on: January 21, 2021, 12:38:31 am »
no. I have contact HP / Agilent but she havent the shematics. So cant buy.