Author Topic: 2215a oscilloscope No trace  (Read 2107 times)
Offline ifrratedTopic starter
« on: January 12, 2016, 04:09:11 pm »
Hi:
I have a 2215A scope with no trace. Beam finder works ok. All trigger settings appear correct.
Printing out service manual today.
Any suggestions as to where to start?  is this a common fail?
Thank you
Offline anachrocomputer
Re: 2215a oscilloscope No trace
« Reply #1 on: January 12, 2016, 04:17:26 pm »
First step, as always, is to check the power supply voltages. A bad or missing power supply voltage will send the CRT beam off the screen, and hence give the symptom that you report.
Offline tautech
« Reply #2 on: January 13, 2016, 07:57:51 am »
Hi:
Thank you
You might need this too:
http://www.sphere.bc.ca/test/tek-parts/troubleshooting-scopes.pdf