Hi all,
I got (yet another for my collection) 2465, it was stated it had problems before I got it, so no problem there.
What have I done so far:
Recapped the Power Supply, cleaned the Siemens fan and powered the scope up.
Power supply is 100% OK now, all voltages are well within range in terms of value and ripple, I set the reference voltage for 10.00V.
Control DAC and reference are checked and set for a 2.5V range.
After Power-On I have a straight trace, not really focussed perfectly yet, but more interesting: I got no proper readout, I have two blurry lines just outside the top and bottom of the screen.
When I apply a calibration signal from the scope itself when on Auto Lvl, the triggered light goes on, but the image looks like if it is not triggered, because I see just two horizontal lines.
They seem to be the high and low level of the calibration signal.
When I switch on the "Voltage cursors" (horizontal lines) they seem OK, a little vague, but their position responds to the delta-settings knobs, but when I switch to the vertical (Time Cursors) I see some vague light in the tube as if the beam is hitting the sides?
The Ch1 trace stays where it is.
Now the weird thing: as I am looking for causes, I tried the following:
When I put the scope on X-Y mode using Ch2 I get a horizontal line instead of a dot??
When I check the outputs of the U800 (second scope) I have clean static DC voltages (with cursors and readout switched off) approx. between 20 and 60V, depending on the Ch1 position, so that should show a dot, but it is a horizontal line......

?
Vertical position of this line (that should be a dot) is nicely controlled by the Ch2 position.
When not in X-Y mode, the horizontal trace appears to be way to wide as well, so it looks like something other than the U800 is causing a horizontal deflection?
Any ideas where this line / horizontal deflection is coming from?
It's a line, so it should be some changing signal, but now I could use some tips and tricks.
Thanks and un saludo,
Leo
P.S.
In my search for the cause I compared signals with my other 2465 and found this so far:
On the A9 High Voltage PCB, in my working 2465 I see a sinus signal on the output of opamp U1890B (Pin7), while on the not working scope I see a negative voltage?