Author Topic: 34401A ACI Fault  (Read 769 times)
Offline WernerFalkenTopic starter
34401A ACI Fault
« on: June 03, 2023, 09:21:36 am »
Good day! I have THREE 34401As with the exact same overarching fault: ACI doesn't work. Tested 6 digit range w/1mA, 100Hz source.
I've tested the following on all three:
  • Self Test: Pass
  • ACV & DCV @ 5V and 50V
  • DCI @ 1mA
  • Freq @ 100Hz
  • Resistance @ 100, 1000, 10,000, and 100,000 ohms
  • Front & Rear Jacks for consistency
  • Checked R120 & R121 for grins
All of the above tests had great results... Tried all the different filters for AC... same issue.
I also tested the resistance across Lo & I input terminals: about 0.65ohm resistance when in ACI (all three of the 34401A)...0.75ohm on a 34461A in 3A range
Looking through the schematics and theory of operation, I am not seeing a clear failure mode for this situation - it's not a relay, as all of the other working modes exercise each of them successfully
All I can think of is that *I* am doing something wrong, but I've checked myself with:
  • 34461A reports ACI @ ~1mA (that's what I am giving it)
  • 3458A reports ACI @ ~1mA
  • Fluke 189 reports ACI @ ~1mA
I know I went overkill on validating my source, but really... I have three bad ones, so might as well prove it's not *ME* three times! I mean, how on earth can I end up with three of these things (from different sellers) with the same symptom?
Any words of wisdom on this?
Thank you!
Re: 34401A ACI Fault
« Reply #1 on: June 03, 2023, 09:34:17 am »
Foul language is coming out of my mouth right now.
Problem solved. It wasn't *ME*, but it was my equipment. My source, apparently, doesn't like 34401As - not 100% why... maybe the 0.65ohm is a deal breaker.
Right before I shut it all down to hit the sack (2:30AM here), I glanced over at my 33522B and thought.... "Hmmmm, I won't really *KNOW* the current level for it w/o doing some math I am not inclined to do right now, but I can see if the 34401As can measure it...."
2Vpp @ 100Hz w/HiZ load gives ~14mA on all three 34401As and my 34461A.
Problem solved. Never mind. Continue on your way... thank you for listening!
Cheers!
« Reply #2 on: June 03, 2023, 09:36:48 am »
Although... 0.65ohm is nowhere near HiZ... still, problem solved.
Offline JohanH
« Reply #3 on: June 03, 2023, 11:31:40 am »
34401A doesn't specify lower than 1% or 10 mA as the minimum current on the 1A AC range (footnote 4 in data sheet).
There is an undocumented 10 mA AC range on which you can measure down to 100 µA AC. This can be turned on over the remote interface: DIAG:POKE 25,0,1
See the following forum post for further details:
https://www.eevblog.com/forum/repair/hp34401a-1ma-calibrator-and-10ma-hidden-range/msg2439456/#msg2439456
« Reply #4 on: June 03, 2023, 04:43:18 pm »
Those darned footnotes!
Thank you!