34401A Beep on/off
« on: January 10, 2021, 02:54:58 pm »
Hi!

I have 2 34401A multimeters.
On both i tried turn off beep function under menu section, but when i turn off beep function and save setting, the beeper beeps despite that if i press any of front panel keys.

What i am doing it wrong?

Thank you!

 

Offline Hrvoje_CROTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: hr
Re: 34401A Beep on/off
« Reply #1 on: January 10, 2021, 07:41:49 pm »
Beeper Control
Normally, the multimeter will emit a tone whenever certain conditions
are met from the front panel. For example, the multimeter will beep
when a stable reading is captured in reading hold. You may want to
disable the front-panel beeper for certain applications.
When you disable the beeper, the multimeter will not emit a tone when:
1) a new minimum or maximum is found in a min–max test.
2) a stable reading is captured in reading hold.
3) a limit is exceeded in a limit test.
4) a forward-biased diode is measured in the diode test function.
Disabling the beeper has no effect on the tone generated when:
1) an error is generated.
2) the continuity threshold is exceeded.
3) you turn off the front-panel menu.
Turning off the beeper does not disable the key click generated when
you press a front-panel key.
The beeper state is stored in non-volatile memory, and does not
change when power has been off or after a remote interface reset.
The beeper is enabled when the multimeter is shipped from the factory

I've found this above.

"Turning off the beeper does not disable the key click generated when
you press a front-panel key." So i assume that everything is ok with my 34401A's?

Thank you!
 