ParkAndFarm:
Great success!.

Replaced the blown U303 ADG1208YRZ Multiplexer along with the 4 highlighted components in the power supply + L1001 inductor [LPR4012-223DMRC] and everything is cherry again! yeah!!
thanks you for the help wraper aiding in making this happen!


**update 06-15** the cause of the destruction was my fault. I had the front lead set tied to one voltage and a back panel high voltage probe under load with its ground clamped to the wrong unit under test. had the high voltage probe under a 20kv and switched to the back panel HV probe without removing the front panel active B+ reading on top of that.

**ran it overnight everything seems to running perfectly yet along with spot on measurements! yeahhh!