Author Topic: 3458 Ram Test 1 Low  (Read 4819 times)
Offline bingo600Topic starter
3458 Ram Test 1 Low
« on: February 01, 2017, 02:23:39 pm »
Well i guess it's my time.
I just got a Ram Test 1 Low , when powering up my 58' , it's been off for 3..4 months.
A power off/on cleared the error , but i suppose the battery has gone low/weak on one of my Dallas ram's.
I just got the 2 data FRAMS from this thead , and i'm waiting for the cal FRAM  (I have a backup of the calram via gpib)
https://xdevs.com/fix/hp3458a_u4/
https://www.eevblog.com/forum/metrology/yes-another-3458a-repair-and-nvram-replacement-to-ramtron-fram/?all
The Ram Test 1 id cured by replacing the data RAMS with FRAMs correct ?
No mods are needed on the SOIC adapters on the data FRAMs correct ?
I'll start looking if i have 2 x SOIC-28 to dip for the data FRAMS , and another to mod for the CAL-Fram.
I'll also check if i have 2 x 28-pin and a 24-pin machined socket in the drawer.
But for now all i loose is any customs settings correct ?
/Bingo
« Last Edit: February 01, 2017, 02:30:11 pm by bingo600 »
Offline TiN
  • Country: ua
    • xDevs.com
Re: 3458 Ram Test 1 Low
« Reply #1 on: February 01, 2017, 04:32:59 pm »
:o You are volt-nut criminal! I always feel bad when 3458 have no life-critical important resistor or LTZ-reference measurement ongoing for a day...  :popcorn:
True.
True.
Correct, hotkeys and settings stored in main RAM.
The following users thanked this post: bingo600
« Reply #2 on: February 01, 2017, 05:00:49 pm »
I usually use the 3457 for doing daily logging stuff , and just the 58' if i need to measure Geller or LTZ
/Bingo
Offline quarks
« Reply #3 on: February 02, 2017, 03:52:07 pm »
in all my tests the F-RAM replacement for DS1230 (I made no mods) worked without any problems, but I have only made quick tests and put back the Dallas Chips afterwards
« Reply #4 on: February 02, 2017, 04:48:20 pm »
Thanx quarks  :-+
I think i'll wait for the CAL-FRAM before installing all 3
/Bingo