So I just purchased a 3478A off FleaBay, and overall the condition of it is good. Did my checks the Battery is good, but the rifa caps on the input power needs to be replaced. Measurements look good, pretty accurate and on point. Was able to verify a couple of ranges with the USA Cal Club references. However as the title shows I'm experiencing a chirping with the unit when I plug in my probes. I'm guessing it's the relay switching into the 30v range when I do. As it's throwing the meter above ±3v. What I'm wondering is why would it be at ±3v with nothing connected? Also why does it only chirp for the 30v range and none of the other ranges? I would assume it should be around 0v not floating at 2.7v nearing 3v, but this is my first big boy meter, and want to be sure. You put any other sort of resistance on it, even say another meter it goes straight down to 23mVDC as I would expect it to be in the first place.
Anyways the other thing I noticed which seemed odd. As I was reading through the 3478A Service Manual it states.
Check and make sure the Reference Supplies are at the correct level and quiet. The +10V supply can be checked at U405 pin6, the -10v supply at U404 pin 6, and the buffered +10V supply at JM201 (See Schematic 3).
However when checked U405 pin6 is -10.23V, U404 pin6 is -10.23V, and JM201 is +10.23V. The Zener diode u461 should show me +7v, I get -7.04v.
Yet later on down in the manual it shows a diagram with U405 pin 6 being -10v, U402 pin 6 being +10v and the zener diode being -7v. Which would all be correct.
Sooo I'm lost. If all seems well then great! If not let me know what you think.
Also I'm looking at these for my rifa cap replacements.
https://www.mouser.com/ProductDetail/EPCOS-TDK/B32022A3223M?qs=%2Fha2pyFaduhDd66L0KefaR5dp143EcXkFI5dzp77CY4P4zuMDC0wwA%3D%3Dhttps://www.mouser.com/ProductDetail/EPCOS-TDK/B32021A3102K000?qs=%2Fha2pyFadughoIb%2FOpAtVzl%2FXOvmVp2GKlY13wf3B92cwoG5csnYQw%3D%3DShould be ok? 2619A is the first few of my serial.