| Electronics > Repair |
| 465B X-Y Horizontal Shift |
| (1/1) |
| awesomechapro:
Hi all, I've recently been restoring an old Tektronix 465B oscilloscope. I've gotten it mostly working, but am having trouble with the X-Y mode. The normal sweep is positioned in the right spot horizontally. (in the center of the horizontal adjustment knob) But when I switch the oscilloscope into X-Y mode the dot is significantly shifted to the left. Is there an adjustment somewhere that can adjust the default horizontal position in the X-Y mode without interfering with the horizontal position of the sweep? I have already had a look throughout the service manual, the calibration procedure and the schematics to no avail. Thanks in advance. |
| tggzzz:
What do you mean by "significantly"? How does the issue change when you change the X-channel control settings? |
| awesomechapro:
as in when both channels a GND coupled, and the horizontal position control is at it's right most position, the dot is still to the left of the center line |
| ondraN:
The X position can be affected by the correct setting of the trimmer R1126 "CH1 TRIG DC BALLANCE" on the CH1 preamp board. (1) |
| Navigation |
| Message Index |