Author Topic: 54622D Boot data error occured?  (Read 1499 times)
Offline Hrvoje_CROTopic starter
  • Country: hr
« on: March 29, 2020, 02:06:54 pm »
Hello,
I have updated few days ago my scope with newest firmware.
Everything was ok until today when i powered it on i've get one new system concern which was not previously seen.
It was boot data error occured.
See the picture
How to fix that? Is this due to low battery?
Otherwise, scope pass all the tests and it is working normally.
Thank you in advance!
Offline JFJ
Re: 54622D Boot data error occured?
« Reply #1 on: March 29, 2020, 03:56:41 pm »
How to fix that?
Replace the lithium cell and reset the oscilloscope:
NB You will need to access the underside of the system board to unsolder the battery.
The following users thanked this post: Hrvoje_CRO
  • Country: hr
« Reply #2 on: March 29, 2020, 05:04:16 pm »
How to fix that?
Hello JFJ,
So if i change the battery and reset the scope, will boot data error dissapear? I am so concerned about that error, i hope it is not HW problem because scope works fine.
Thank you!
Offline JFJ
« Reply #3 on: March 29, 2020, 05:56:45 pm »
You should be able to test that the errors will disappear by doing a power-on reset now. The boot data that was lost is only non-critical information, such as the oscilloscope's settings (Volts/div, Time/div, etc.) before it was last powered down.
« Last Edit: March 29, 2020, 06:04:47 pm by JFJ »
  • Country: hr
« Reply #4 on: March 29, 2020, 06:06:12 pm »
I tried that! Pushed softkey during power up and display came wihout any errors.
So this is a good news?
Yuk! Battery replacement is sensitive but i will change it.
  • Country: hr
« Reply #5 on: March 30, 2020, 01:01:16 pm »
I forgot to mention, i did as you wrote and errors did disappear, but at the second power-up it is the same error list.
Thank you!