No signal in, the horizontal is using the magnifier to i can see the problem better. I think i'll start with checking the power supply
If you set the timebase to line triggering, then you will be able to see if this is from AC line ripple or something else. These oscilloscopes are old enough that it is likely the aluminum electrolytic capacitors in the power supply are worn out.
To add to the crt image problem, the trace is also bending slightly to the corners..so after i check the PS i'll try and go throug the crt calibration. What do you think?
The CRT calibration may affect this but it is not serious. CRTs usually have some geometry distortion at their edges.
There are actually *two* adjustments which affect CRT geometry. R41171 adjusted the CRT geometry but the vertical deflection plate shield voltage adjusted by R41176 also has an effect on CRT geometry.
Vertical in the horizontal slot: 3V square calibrator from a 1054z is exactly 5 divisions at 500 mV/div
So the lack of horizontal deflection is probably what I suggested. Check the gain potentiometer (R4310) on the horizontal CRT amplifier first. It may be bad or a previous owner may have misadjusted it.
Low post-deflection acceleration voltage?
This would affect both horizontal and vertical deflection. Vertical deflection is apparently accurate.