Author Topic: HP 8563E repair  (Read 1119 times)

0 Members and 1 Guest are viewing this topic.

Offline EE4allTopic starter

  • Regular Contributor
  • *
  • Posts: 130
  • Country: us
HP 8563E repair
« on: December 21, 2022, 08:52:06 pm »
I am working on a 8563e that shows error 334 LO AMPL. The YTO is fine, and I can show signals on the trace, but they appear low, especially in the low band. Checking the LO distribution amplifier calibration,  I can't even adjust the voltage into the specified negative range. Likely a bad amp, or the drive control for it.
« Last Edit: December 27, 2022, 04:46:22 am by EE4all »
 

Offline EE4allTopic starter

  • Regular Contributor
  • *
  • Posts: 130
  • Country: us
Re: 8563E repair
« Reply #1 on: December 21, 2022, 11:08:02 pm »
Ok, CLIPs and service manual in hand. I will follow the diagnostic procedures and report back.
 

Offline EE4allTopic starter

  • Regular Contributor
  • *
  • Posts: 130
  • Country: us
Re: 8563E repair
« Reply #2 on: December 27, 2022, 04:45:18 am »
Just to close this off, the YTO was working, but output was far too low for the amp to work properly without throwing errors. I replaced the YTO with a strong one, and the LO error went away.
 

Online Tony_G

  • Frequent Contributor
  • **
  • Posts: 964
  • Country: us
  • Checkout my old test gear channel (link in sig)
    • TGSoapbox
Re: HP 8563E repair
« Reply #3 on: December 27, 2022, 05:22:35 am »
Just out of interest, did you get a YTO from another 8563E (or one that was a spare part) or did you use a different YTO and work it to fit?

I have an 8563E that works fine but just interested in what you did to repair.

TonyG

Offline EE4allTopic starter

  • Regular Contributor
  • *
  • Posts: 130
  • Country: us
Re: HP 8563E repair
« Reply #4 on: December 27, 2022, 05:38:46 am »
I got the YTO from another 8563E that I got for parts. The original YTO was completely dead, and I ordered a replacement refurbished one off of ebay. That is why in my first post I assumed the YTO was ok, as all the errors except the LO AMPL one went away after making the adjustments for the new YTO. I would later figure out that although it worked, the output being under 0 dBm was far too low. A few days later another very old and broken 8563E that I bought showed up, and it had a strong YTO! I was able to use it to fix the first 8563E. The refurbished YTO is going back to California.
« Last Edit: December 27, 2022, 05:40:44 am by EE4all »
 

Offline EE4allTopic starter

  • Regular Contributor
  • *
  • Posts: 130
  • Country: us
Re: HP 8563E repair
« Reply #5 on: December 27, 2022, 04:40:27 pm »
Here is how I was measuring the YTO output power level, as I didn't have an analyzer that would cover such a high frequency. I didn't bother dialing in the offset to the power meter (for the attenuators). There are 40 dB worth there, as that sensor has a maximum input of -20. Setting span to 0, and then measuring it, the highest point was here, with -4 dB.

« Last Edit: December 28, 2022, 12:48:05 am by EE4all »
 