Hello everyone,
I have recently acquired an HP 8568A and his 85662 IF/Display. As usual with AS-IS item/Found in a trash item, they are not working.
I have no CRT activity and both INSTR CHECK LED off, keyboard inoperative after start up, so not an easy failure.
What i did so far:
Checked both RF/IF Section supplies - OK
Checked 50-wire interconnect cable - OK
Checked keyboard while startup - OK
Ran HP-IB Test - OK
Ran LONG-POP Test - OK
Ran Digital storage Test - NOK, No test pattern but INSTR CHECK LED left on, as expected.
Hooked an external XY monitor - OK but it seems to lack some data
Checked CRT Filament resistance - 19 Ohms, OK
Checked CRT Voltage on A1A3TP1AB - 0V , NOK
Checked +18V on A1A6TP100-5 - +18.12V, OK
Checked DRIVE on A1A6TP100-4 - No oscillation, NOK
Checked CRT DSBL at 0V on A1A6 - 4.5V, CRT is disabled, NOK
Visual check of A4A9 Board and U6 in charge of CRT DSBL signal - OK
Now i can say my HV Power supply and my CRT are good, they are just disabled by the RF Section but why ?
Like i said display is working on an external XY display, but it lacks quite a lot of data, see attachments.
Have anyone encountered this problem before ?
In the waiting for your kind answers, i'm going back to the service manual !
Best regards,
May the force be with you.