Hello,
Thank you for your quick reply. In fact, I don't think I have a problem with the ref. generator, I checked the output of the reference signal (image attached) and it seems correct.
I have seen the following procedure in the manual but I wonder about the flashing error message, it seems to me that the procedure talks about permanent failure.
Also, the description of the measurements doesn't seem clear in the doc.
perhaps it's a known fault?
The battery is old and should be changed, but I don't think it's the cause of the problem.
1. Check the rear-panel cable connections for the 10 MHz reference.
When using either the standard 10 MHz timebase or the precision
frequency reference, make sure that the 10 MHz REF OUT on the
rear panel is connected to the EXT REF IN. The W6 jumper is
supplied with the analyzer for this purpose.
When using an external 10 MHz reference, make sure it is connected
to the EXT REF IN on the rear panel.
2. If the rear panel connection is correct, check the power output of the
10 MHz frequency reference.
• The A22 standard 10 MHz timebase (RTXO) has an typical
output of 0 dBm ± 6 dB.
• A22 precision frequency reference (OCXO), has a typical output of
0 dBm ± 3 dB.
• An external 10 MHz reference should have an output between
-2 dBm and +10 dBm.
3. If the output of the 10 MHz reference is correct, check the continuity
of W18, the EXT REF IN cable, and W17, the 10 MHz OUT cable.
4. Check the continuity of the VTO_TUNE control line.
5. Check the continuity of the W14 ribbon cable and the W21 coaxial
cable.
6. Check the continuity of the A25 assembly connections to the A9
assembly that pass through the A15 motherboard.
7. If the error message is still present, suspect the A25 counter lock
assembly.