https://www.artisantg.com/info/HP_8904A_Service_Manual.pdfPage 135 has the circuit
I have two of these units, when floated one unit measures some megaohm to ground, the other unit measures 1.000 K to ground. It looks like its measuring the R423 resistor behind the CM choke, while the other one is not, at least thats my hunch. Both boards look the same on thermal and it looks like the correct relays are getting warm on float switching.
I am probobly missing something simple, but why is there such a big difference in input impedance? Both units have 2 output cards, so 3/4 cards measure 1 megaohm, while 1/4 cards measure 1k.
If I set it on a 5Vpp output and measure sine wave while its floating, 3/4 channels measure 1 megaohm and the other channel reads 300 ohms to earthground (dancing around), but the output voltage is correct
where should I look?
so in short
1) channel 2 on unit #2 has a 100x lower isolation impedance when measured with DC ohmmeter
2) isolation impedance on channel 2 unit #2 has a isolation voltage that seems dependent on output voltage setting of output amplifier