Yes, I fixed it. As the RBW decreases, the band of the filter in the picture above becomes narrower and narrower. It turned out that the resonance was shifted away from the required 21.42 MHz.
After a detailed inspection, it turned out that the adjusting screw of the L14 coil was broken off - someone tried to adjust the coil and the plastic screw broke off. There are several similar coils in the device, avoid adjusting them, the plastic screws have become very fragile. If you still have to adjust, first carefully dissolve the fixing glue on top of the coil.
In my case, it turned out that the L13 coil was broken and because of this, the resonant frequency of this block shifted. Someone tried to configure it using the L14 adjustment and it is possible, but not correct. Start by checking L13 and replacing it if necessary. It is important to comply with the DCR for L13. On this board in another place there is the same coil and it seems that the DCR was 20 ohms, at first I bought a coil where the DCR was 7 ohms - it is impossible to perform auto-calibration with it.
To check, you need to connect a network analyzer to this board, you can find this in the instructions in the calibration section. You need to get the result below:
(before and fixed - the center must be 21.42MHz)