Hello,
Recently I almost finished working on Advantest R3271A 100Hz-26GHz spectrum analyzer. I've got this one cheaply as nonoperational / unknown condition. This is rather old instrument from Japan, with rather nice parameters - especially spectrum coverage. Also all service manuals with schematics are available online.
The unit was dead, when powering it on I could see a led blink on a front panel, but nothing else. It also turned out that someone already worked on my unit, many screws were missing, front panel was not fitted right to the frame and so on. I of course check the power supply - no voltage at any of the outputs.
To get to the power supply almost everything needs to be disassembled (I wanted to this anyway so no a big deal). Power supply itself was not made by Advantest and it was incredibly complicated (2 boards, 8 transformers/inductors, tenths of capacitors). Potential problem could be seen at a first glance - corrosion near capacitors. ABSOLUTELY EVERY electrolytic capacitor in the PS leaked.
Replacing electrolytic caps fixed the PS and the unit started up to its normal screen, but the displayed spectrum was not right - off by couple of MHz and many additional tones near the main were displayed. The problem was 3.81GHz PLL for the second mixer, that is used in lowest band only. One SMD ceramic capacitor in phase detector become a resistor (or residue under it) and it was dragging output to VCO down. Replacing it fixed the unit. I also fix broken screen cover, and assembly everything right.
Now there is one more thing - selecting frequency span larger then 4GHz does not work - unit shows wrong frequency, and wrong amplitude of the peak. I already know that the circuit driving main YTO is faulty. Advantest generates the frequency sweep by analog circuits, and >4GHz range is the largest. I'm guessing that one of the analog switches in the line is faulty (possibly has a partial short to GND), and when activated in the >4GHz band drags the sweep signal down a bit. That signal is then input to main YTO (unit 4-8GHz LO) driver and we end up with the wrong frequency. I am waiting for the parts now.
Last thing is burned out screen, it is still visible, but poorly. I have a 640x480 LCD panel with the right dimensions, but it is a bare panel. I don't know if creating a driver for it (using STM32F4 with integrated display driver) is a good idea.... It would require taking pixel data from the Advantest at a component level - there is only primitive composite output provided.
I hope that someone will find this description and pictures useful,
Michael Widlok