Yesterday I now also fully calibrated my Tektronix 2440 scope externally!
And because my Tektronix 2440 scope repair now also was FRAM safe and fully calibrated I posted it after my last 2440 scope article on Jestine Yong's blog.
https://jestineyong.com/tektronix-2440-oscilloscope-repaired-part-4/ I first upgraded my oscilloscope by replacing both old battery backupped 32k SRAMs U664 and U350 by FerroMagnetic FM1808 DIP 28pin chips. Because without the calibration data in these chips the scope won't work, no working screen, no nothing, so I also used the uploaded firmware bin files from this great guy - see following link - to keep my 2440 alive after replacing both SRAMs. And pre-programmed my FM1808 with them so my scope would still work to be able to calibrate it again afterwards! Those bin files (from an almost identical 2430 oscilloscope) can be found here:
https://www.diyaudio.com/forums/equipment-and-tools/312894-tektronix-2430a-lost-cal-data-repair.html . And if you have a dead 2440 most likely these SRAMs lost there original firmware and need to be restored first! PS: I used my even almost as old SYSTRON 100A Pulse Generator to finish the last external CTE calibration check. And I now have a fully Internally (Selftest) and externally calibrated Tektronix 2440 oscilloscope. Also the above given link provides the easy Calibration steps you can follow.
I also read some guy had problems with calibrating the attenuator at 0.2V, 2V , 20V on Channels 1 and 2. I had the same problem but noticed when checking the voltages with my DMM that only a voltage of about 1.999 V passed. And 2.0 was too much which failed passing.
And in Part7 of this Youtube video you can see how to setup the signal for the final CTE calibration step: