Hi!
Before I start with the questions, let me introduce myself.
I'm Dave, and engineer working for the University of Twente and I'm quite ok in repairing equipment.
Now I'm having quite some trouble with one of our old oscilloscopes.
I'll start with describing what DOES work, and later on tell what I've done to repair it.
So when the power of the scope is turned on, the lights of the button/knobs panel blink once, and afterwards the 'More' button keeps lit.
This is all. With another scope (54622A) this happens as well, but then it starts 'booting' up.
I have taken apart the scope, and started measuring the voltages as described by the repair manual.
These voltages should be: 5.1V, -5.2V, +15.7V and +3.4V.
Now all these voltages match, except for the -5.2V. This was around -8V. According to the manual I should already replace the power supply, but as this thing is 1000 years old, getting by one might not be easy. So what I did was to replace (almost) all aluminium capacitors and try again.
I have not yet retested the voltages, as I;m not quite sure that this single voltage would mess it all up. Does someone else have the same experience with this machine? Or am I missing some important stepts? This is the first scope that I'm trying to repair, but luck has not yet been on my side.
For the people still reading, thanks for doing so, and I hop you might have some insight on how to repair this ancient device.
Dave