I recently bought a Agilent 34401a from Ebay. I replaced all electrolytics and tantalums, all works great but the AC current bellow 10 ma AC, it was the same before replacing the caps and re-calibrating the 10 ma AC as per the calibration procedure. So at 10 ma AC and above it works normal and the readings match closely with a Keithley 2001 and a Siglent SDM3065X that I have, bellow 10 ma AC it rolls off, meaning 5 ma going in reads 2 ma, 1 ma going in reads 23 micro amps AC. AC volts, DC volts, DC amps and resistance all works fine. Any help would be appreciated.
Thank you.