Author Topic: Agilent 34401A Errors 612/613, 615 and 617-619  (Read 1198 times)
Offline OleTopic starter
« on: June 09, 2022, 10:05:14 am »
I have bought a used 34401A.
The Basic Rundown:
-Voltage Measurements via Input is working
-Current Measurement is Working
-Front Connectors have been used
-Ohms Measurement does not work
-Meter is from 2006 with 0609 on LM399
-Big Electrolytics seemed fine for a smoke test
The Selftest gives the following Errors:
612: Ohms 500 nA source failed
613: Ohms 5 uA source failed
615: Ohms 10 uA source failed
617: Ohms 100 uA source failed
618: DC high voltage attenuator failed
619: Ohms 1 mA source failed
I have not yet desoldered any components
*record scratch noise* Hey, you.
Yes, you. Have an awesome day!
Offline strawberry
  • Country: lv
Re: Agilent 34401A Errors 612/613, 615 and 617-619
« Reply #1 on: June 09, 2022, 10:22:34 am »
Q202 Q211
Online Kleinstein
« Reply #2 on: June 09, 2022, 10:26:09 am »
The 2 JFETs from the ohms part are prime canditates to check.
The error 618 about the divider may also be caused by the ohms current source not working. So I would fix the ohms part first and this may also fix this error.
Offline HighVoltage
« Reply #3 on: June 09, 2022, 10:26:46 am »
Also, use the search function, we have discussed these errors a few times and you can learn lots from these posts. Like this one
https://www.eevblog.com/forum/repair/agilent-34401a-error-606/msg934953/#msg934953
« Reply #4 on: June 09, 2022, 11:31:32 am »
Q202 Q211
Luckily those are TO-92s so I can desolder them easily.
I dont have the full parts list on hand, so I might need some help for finding the replacement parts.
Q202: [1671  (8)08] is written on that one. Not sure about the first eight in the second block
Q211: [BF47  PN  4117A] not 100% sure might be a simple 2N4117A.
The big Elkos in mine are all:
a) good looking
b) non bulging
c) covered with a piece of kapton (?) tape on top
C752: 4700u 25V 105°C
C556: 2200u 25V 105°C
C553: 1000u 50V 105°C
C551: 1000u 50V 105°C
Offline HighVoltage
« Reply #5 on: June 09, 2022, 12:10:47 pm »
The big capacitors on these instruments usually never go bad.
But make sure to test all rail voltages and ripple on them.
« Reply #6 on: June 09, 2022, 01:05:03 pm »
Well, if those Caps didnt look like freshly soldered in I would have switched them for fresh ones before giving the instrument a smoke test. Now only one question remains: What kind of JFET´s are Q202 and Q211?
Offline bdunham7
« Reply #7 on: June 09, 2022, 01:39:57 pm »
These are TO-92?  According to the Agilent manual and my recollection, I thought they were SMD.  In any case, if they are TO-92:
Q211 is a PN4117A and is generally in stock at places like Mouser.
 Q202 is a selected 2N5461.  I do not know the selection criteria, but the basic 2N5461 is still available from Jameco, although it is listed as obsolete/unavailable elsewhere.
Offline Dr. Frank
« Reply #8 on: June 09, 2022, 03:02:14 pm »
Hello Ole,
that's a common problem, better see this link: https://www.eevblog.com/forum/repair/hp-34401a-error-612-613-615-617-618-619-621/75/
In most cases it's U201, an AD706, or it's a relay.
First, check the values of driving currents when in 4W Ohm.
Frank
« Reply #9 on: June 09, 2022, 03:32:33 pm »
Q211 i will order a few replacement PN4117A from Mouser and as for the Q202, the 2N5461 is availible for cheap at a online shop, alternatively I thought about purchasing a few J176/J177 P-Channel-JFETs and replacing the Q202 with one of those.
Offline strawberry
  • Country: lv
« Reply #10 on: June 09, 2022, 05:58:35 pm »
could check first with multimeter if it is blown or not