Picked up a used 34401a off ebay. The unit passes self test without errors.
RS232 works but GPIB does not.
Is there a common cause of failure for 34401a GPIB ports?
I'm guessing either U703 (75ALS162) or U702 (75ALS160) is bad. Is there any way to tell which one?
« Last Edit: April 23, 2017, 03:16:56 am by usagi »
Online Dr. Frank
Re: Agilent 34401a repair - dead GPIB
« Reply #1 on: April 18, 2017, 09:56:49 am »
Hello usagi,
see manual p. 92: 'Only one interface can be enabled at a time'.
So I assume, that you have to enable 'HP-IB / 488' first, in the I/O menu, see p. 162, and then it should work.
Frank
« Last Edit: April 18, 2017, 10:00:03 am by Dr. Frank »
« Reply #2 on: April 18, 2017, 05:11:57 pm »
Hello usagi,
Frank
i know this already  :) i had to go into those menus to enable RS232 in the first place, since it came with GPIB interface enabled by default.
so I can confirm GPIB is 100% busted.  :scared:  GPIB works on my other devices - keithley 2015 and R&S cmu200.
so do you know which exact 34401a IC to verify for GPIB troubleshooting?
« Last Edit: April 18, 2017, 05:18:32 pm by usagi »
Offline Tony_G
  • Checkout my old test gear channel (link in sig)
    • TGSoapbox
« Reply #3 on: April 19, 2017, 08:38:01 pm »
I just repaired an 11713A that didn't work with GPIB by resoldering the GPIB connector - Might be worth giving it a try if the connector is soldered direct to the board in those devices.
« Reply #4 on: April 20, 2017, 07:24:27 am »
100% confirmed the connector is perfect.
100% confirmed an IC is blown. waiting on mouser order to arrive now.
Online HighVoltage
« Reply #5 on: April 20, 2017, 07:39:32 am »
Once you exchange the GPIB controller it should work perfect again.
« Reply #6 on: April 20, 2017, 08:25:08 am »
U701 is fine.
Online HighVoltage
« Reply #7 on: April 20, 2017, 08:44:40 am »
U701 is fine.
Ahh, which IC do you think is blown?
« Reply #8 on: April 23, 2017, 03:16:40 am »
replaced blown IC, GPIB now works. again :-DMM