First of, I apologize for asking for help as my first question. I have been reading the forums for many years and never realized that I did not sign up for a log-in. I very rarely ask questions, I just soak up knowledge by reading and keeping my mouth shut.
Anyway...
I acquired an Agilent 34401A recently and am stumped by it. Disclaimer: I fix audio gear but I am not a "test gear" guy, so while I can understand guidance and use it, I am not well versed in how digital meters actually work.
So the problem is: Error 611, 612, 615, 617, 619. When I cold start the meter and run self test Error 612 was NOT there, but comes back shortly.
The meter reads DC, AC, OHMS and CURRENT. But all ranges and readings are a little low, by the same amount.
For example: 5vdc will show 05.00047 on my good meter and 04.99609 and bouncing the last 2 digits. This is consistent across ALL ranges as if the reference is bad, but it measures 7v and +10/-10 at the respective points. Hmmm...

Power supply voltages measure within spec, and U201 was previously replaced. As far as I can tell, all the Zeners are OK, as I can measure the correct voltage drop across them. All caps APPEAR ok, and ripple is a few millivolts across each.
I DID need to clean the FRONT/REAR switch, as it was twitchy. OK now.
So.. I'm a bit stumped. I read everything I could find and checked what I could. I can't seem to find a definitive fault.... I guess neither could the guy that sold it. Upside is that it's a clean meter with an excellent display and it was cheap enough to take a chance. The seller DID disclose all the error codes, but I bit off more than I could chew. In the words of Jeremy Clarkson... "How hard could it be?"
Can anyone give me some insight as to where to look next?
And as you can see I have a working and spot-on 34401A bought because of YOU guys and am very happy with it. Was hoping for a second one without giving a kidney!