This 34401A was working perfectly until yesterday.
Turning it on today, gave an "ERROR" message in the front display after bootup.
Interestingly, the self test is passing without any problems.
Since the 34401A stores only up to 20 Errors, there are probably more, but may be they era just stored new, every time the unit is turned on.
Following Errors are listed in the Error section of the Menu:
740, 747, 748, 741, 742, 743, 744, 745, 746, 740, 747, 748, 741, 742, 743, 744, 745, 746, 740, -350
The last (-350) one also indicates that there are more than 20 errors.
The user guide is listing these errors as:
- 740 Cal checksum failed, secure state
- 741 Cal checksum failed, string data
- 742 Cal checksum failed, DCV corrections
- 743 Cal checksum failed, DCI corrections
- 744 Cal checksum failed, RES corrections
- 745 Cal checksum failed, FRES corrections
- 746 Cal checksum failed, AC corrections
- 747 Cal checksum failed, GPIB address
- 748 Cal checksum failed, internal data
Does anyone here have an idea where to start to look for the problem?
Thanks for any advise.