Agilent 34401A Serial port pints 3 (TX) 5(GND) stuck at 9V on sys (SOLVED)
The_Spectrum.A_idiot:
Dear All ,
After finally fixing my instrument I have a new issue.
THe serial port regardless of its selected on the I/O between RS232 or GPIB the device directly applies 9V on the TX pin.
Is this normal ? or I am looking at another issue ?
Thanks
Addendum :
Using wrong method to talk to the instrument
macboy:
RS232 idles at the high voltage level, so this is normal.
squadchannel:
RS-232 protocol, it would be normal operation. There is no problem.
RS-232 detects "0" at positive 3-25V and "1" at nagative 3-25V, not TTL.
The_Spectrum.A_idiot:
okay , but when I connect a USB to serial device seems to not be happy with said levels.
Do I need so have something else in-between the instrument and the USB to serial device ?
Is the first time I am encountering such issue.
jchw4:
--- Quote from: The_Spectrum.A_idiot on October 03, 2024, 11:29:48 am ---okay , but when I connect a USB to serial device seems to not be happy with said levels.
Do I need so have something else in-between the instrument and the USB to serial device ?
Is the first time I am encountering such issue.
--- End quote ---
https://www.keysight.com/us/en/lib/resources/training-materials/communicating-with-instruments-using-rs232.html
TLDR: you need correct cable.