In this episode Shahriar repairs a faulty Agilent 3458A. These highly desirable multimeters have continued to climb in price in the past few years. The last time The Signal Path has repaired one of these units was back in 2018.
This instrument generates several errors both during power up as well as Auto Cal procedure. The error points to a failure in the OHMS measurement mode. Detailed instrument block diagram and schematics are examined. Measurements also show incorrect DC current sourcing during OHMS measurements and the associated circuits are closely analyzed. The problem is traced to a quad-comparator IC. Replacement corrects all problems during ACAL. The battery backup of the SRAM ICs are also replaced to eliminate non-volatile errors. The IC of the comparator chip is also extracted and examined under the microscope.
The repaired instrument performance is verified by using voltage, current and resistance standards.