Electronics => Repair => Topic started by: HighVoltage on December 11, 2015, 05:39:28 pm

Title: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 11, 2015, 05:39:28 pm
This Agilent 34970A Data Acquisition unit I bought used in broken condition very cheap.
After opening it up, I found the problem right away with a loose ribbon cable.
Then I took it completely apart and cleaned it all and replaced a ribbon cable.

It seems to work perfect now.
But I measured the battery and it has only 2.6V instead of 3V
So, I would think it needs to be replaced.

Does anyone here have experience with replacing this battery?
Do I need to keep the unit powered up or at least plugged in, during the exchange of the battery.
Anything else to watch out for?

Here are some pictures:
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: nfmax on December 11, 2015, 05:51:04 pm
Yes, I have replaced the battery in the unit I bought second-hand a few weeks ago. You should disconnect the unit from the mains supply, calibration data (including relay operation cycle counts) is not dependent on the battery. The procedure is described in the service manual and it's pretty straightforward.
My unit had a cylindrical, through-hole tag mounted battery, not a surface mount button cell like yours. I think these have a larger capacity: if you look at the PCB you will see there are three holes for the battery tags so presumably you can use either. Be careful to choose a battery with the correct pin spacing at the 2-pin end, and with the correct polarity - all possible variants are made and only 1 of the 4 is right!
It sounds as though your battery, while dying, is not quite dead yet. When it finally dies, you will get an error indication each time you power up the unit & 2 codes - clock lost & memory lost. But no harm in replacing the battery now while you are working on it anyway!
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: magnehh on December 11, 2015, 10:02:17 pm
I believe the HP 34790A's had the "long" cell and the old, horrid FRAM. While in the Agilent version, they changed both of them.

I do have two multimeter modules for these if you're interested. Only problem is, they loose the cal data when you unplug them from the main board. :-P
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 12, 2015, 12:02:45 pm
Thank you both for the feedback.
Now I just have to find the right battery with the solder tabs at the right spot.

The original Panasonic number is BR2477A
That one seems not to be available.



Title: Re: Agilent 34970A, Battery replacement and repair
Post by: PA0PBZ on December 12, 2015, 12:26:04 pm
http://www.digikey.com/product-detail/en/BR-2477A%2FVAN/P030-ND/273650 (http://www.digikey.com/product-detail/en/BR-2477A%2FVAN/P030-ND/273650)
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 12, 2015, 12:58:43 pm
Thank you for the link, PA0PBZ
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: nfmax on December 12, 2015, 01:21:19 pm
Thank you both for the feedback.
Now I just have to find the right battery with the solder tabs at the right spot.

The original Panasonic number is BR2477A
That one seems not to be available.
The battery I fitted was a Varta, Farnell part number 197-2505. This has two pins at the positive end, spaced at 7.5mm. The dead battery was reading about 0.65V (!), and I was seeing error codes 202 and 204 at each power on. I don't think this was the original battery though as there was uncleaned flux on the PCB. My unit is an Agilent labelled one, as are the plug in modules.
I have a spare copy of the (English) service guide - PM me if you are interested

Max
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 12, 2015, 02:02:48 pm
PM sent,
I ordered the original battery from Digikey and I will post again, once the replacement battery is installed.
May be a full calibration will be required afterwards.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: Jwalling on December 13, 2015, 12:10:51 pm
PM sent,
I ordered the original battery from Digikey and I will post again, once the replacement battery is installed.
May be a full calibration will be required afterwards.

FYI: I've found that these units will accept a standard horizontally mounted CR2032 battery holder like you would see in any PC motherboard. Sure the battery doesn't last as long, but when you need to replace it, it's a snap.

Jay
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 13, 2015, 12:33:35 pm
FYI: I've found that these units will accept a standard horizontally mounted CR2032 battery holder like you would see in any PC motherboard. Sure the battery doesn't last as long, but when you need to replace it, it's a snap.

Jay
Thank you for this information, great idea!
I will get a CR2032 holder and take a closer look.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: Jwalling on December 13, 2015, 01:35:09 pm
FYI: I've found that these units will accept a standard horizontally mounted CR2032 battery holder like you would see in any PC motherboard. Sure the battery doesn't last as long, but when you need to replace it, it's a snap.

Jay
Thank you for this information, great idea!
I will get a CR2032 holder and take a closer look.

You're welcome. Note that the pins of the holder don't go in the same holes as the original battery, they go in the larger holes between the original pin holes. You'll figure it out using a DMM on continuity - just watch out for polarity. You'll also need to scrape away a bit of the solder mask before soldering it in. I use a bit of scotch tape to hold in the battery holder since you have to solder it in upside down.

Jay
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: dom0 on December 13, 2015, 02:55:44 pm
Yes, I have replaced the battery in the unit I bought second-hand a few weeks ago. You should disconnect the unit from the mains supply, calibration data (including relay operation cycle counts) is not dependent on the battery.

The battery powers only the RTC?

In my olde 3478A there was one of these large (thicker but shorter than a AA mignon) lithium cells, over 30 years old, still had 2.95 V or so on it :-)
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: nfmax on December 14, 2015, 11:14:15 am
The battery powers only the RTC?

In my olde 3478A there was one of these large (thicker but shorter than a AA mignon) lithium cells, over 30 years old, still had 2.95 V or so on it :-)
The battery also supplies the SRAM used to hold the measured data - this is really useful as you can take measurements on an experiment in the lab, then bring the datalogger back to your desk to download & analyse the data. The one I bought recently was for home use, but we have used dozens of these dataloggers at work since I bought us the first one back in 1999!
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 23, 2015, 09:12:46 pm
Today I installed the new battery (original replacement from Digikey)
The Battery was out for about 6 minutes and the instrument kept all calibration data.

Tomorrow I will check the calibration on all ranges.
It seems so far, that it was a success.

Thanks for all your help.


Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on December 31, 2015, 05:51:57 pm
I know that it is little bit late, but in my unit fir battery CR14250, it is rotated by 90 degree compare to your picture.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 31, 2015, 05:54:44 pm
I know that it is little bit late, but in my unit fir battery CR14250, it is rotated by 90 degree compare to your picture.
How can it be rotated 90 degree?
You mean 180 degree?
Can you take a picture of what you have in your unit?
Thanks
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on December 31, 2015, 05:59:11 pm
BTW, this 34970A has been turned on for a while now and seems to be very stable.
I still need to do a detailed calibration check.

Do you usually check the calibration on all channels of a 901 card to also include the multiplexer card ?

Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on January 01, 2016, 01:03:57 pm
I calibrate only one channel, but before I highly suggest to check the relay card resistance and relay odometer.

Quite funny is the fan placement on side completely covered with enclosure, I'm quite curious what is purpose of this fan, because on another side it is covered by IEC connector assembly.
So It cannot cannot help to flow air inside instrument, like we saw inside Keithley 7510.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on January 03, 2016, 12:59:16 pm
What I forgot to mentioned - check the relay counter on internal DMM (if present).
I do not find the way how to check it through front panel, only remotely by DIAG:DMM:CYCLES?
For instance on my unit the DMM relay K102 is almost dead >10M cycles. Probably someone selected different ranges 10V/100V or similar switching problem :palm:
For details check "Chapter 5 Theory of Operation - Internal DMM" in service manual.
I needs to desolder relay and replace it by new one Omron G6SU-2-DC3
Because my unit has only one issue ( errors in self test 740/741/748 and etc) all were due to unlocked instrument, when I locked calibration all are gone.Battery has 3V.
Based on the relay counter I expect that unit failed calibration due to K102 relay.

There is another trap - if you checking basic performance of modules put into terminal some wire. Unit has been supplied with two modules. Both as faulty ones with marking which relay is faulty.
So I checked the odometer and maximum was 25 cycles.
DIAG:RELAY:CYCLES? (@101,102, ....)
So I checked them according to service manual and yes they seems to be not working.
So I expect that someone reset relay counter, but modules were almost new. When I measured the relay contacts all were OK, yes the problem was in terminals!!!!! :palm:
The guy which has been cheeking the modules make same mistake probably.
If you are measuring the screw, there is ope circuit, when you put there some wire it is OK :)
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on January 03, 2016, 01:15:50 pm
Thank you plesa, this was really good information.

One more question,
If you would replace a relay, can you reset the counter to reflect the new relay installation?
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on January 03, 2016, 01:30:31 pm
Yes, both can be cleared/reset. You needs to unsecure the instrument
CAL:SECURE:STATE OFF,HP034970

and then
DIAG:DMM:CYCLES:CLEAR  (1,2,3 ) depends on which relay counter you want to reset
DIAG:RELAY:CYCLES:CLEAR (@101,102, ...) depends on which relay counter you want to reset

The good document for this is manual for 34972A instead of 34970A manual
http://cp.literature.agilent.com/litweb/pdf/34972-90001.pdf (http://cp.literature.agilent.com/litweb/pdf/34972-90001.pdf)
or old HP manual
http://www.keysight.com/upload/cmc_upload/All/34970-90003_users.pdf (http://www.keysight.com/upload/cmc_upload/All/34970-90003_users.pdf)

That's pity that this is not possible for bench DMM, or I do not know how.
This can be helpful to predict the relay failure like you have on 34410A.

Which firmware you have and what is response to SYSTem:VERSion?

In meantime I found only service note to my serial number http://literature.cdn.keysight.com/litweb/pdf/34970A-13.pdf (http://literature.cdn.keysight.com/litweb/pdf/34970A-13.pdf) related to stability of 400k Vishay UXB thin film resistor which can affects 10M and 100M Ohms measurement.
And another related to connection bwtween main board and DMM http://cp.literature.agilent.com/litweb/pdf/34970A-14C.pdf (http://cp.literature.agilent.com/litweb/pdf/34970A-14C.pdf)
Funny is that my serial number is not affected by cover exchange, but in real it does not have proper cover ( so check if your cover has ventholes on righ side and bottom end).


Complete list you can found here:
http://servicenotes.literature.keysight.com/litapp/SearchSN.do?method=openExternalSNSearch&prodNum=34970A (http://servicenotes.literature.keysight.com/litapp/SearchSN.do?method=openExternalSNSearch&prodNum=34970A)

Also interesting document about change FRAM ( when you have 740 errors ) in attachment.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on January 03, 2016, 10:54:59 pm
I just found that latest firmware is v13, programmed in U401 ( 27C4096 OTP EPROM). I will really appreciate if someone can post version v12 or v13 for upgrade.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on January 04, 2016, 02:05:30 pm
I have two of these Agilent 34970A units.
How do I test what FW version is installed?
Can I do it over the UI?
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on January 04, 2016, 05:50:35 pm
I have two of these Agilent 34970A units.
How do I test what FW version is installed?
Can I do it over the UI?

Version of FW is in response to *IDN? command, it is first number ( XX-X-X).
Do you have PLCC 44 package programmer?
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: nfmax on January 05, 2016, 02:40:30 pm
I have two of these Agilent 34970A units.
How do I test what FW version is installed?
Can I do it over the UI?

Version of FW is in response to *IDN? command, it is first number ( XX-X-X).
Do you have PLCC 44 package programmer?
Keysight Connection Expert also shows the firmware version from the ident string. My unit returns 12-1-2. Does anyone know what the differences are?

Max
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: Jwalling on January 05, 2016, 02:49:05 pm
I have two of these Agilent 34970A units.
How do I test what FW version is installed?
Can I do it over the UI?

Version of FW is in response to *IDN? command, it is first number ( XX-X-X).
Do you have PLCC 44 package programmer?
Keysight Connection Expert also shows the firmware version from the ident string. My unit returns 12-1-2. Does anyone know what the differences are?

Max

Take a look at service notes - there's a lot of info on the firmware (and other) changes:
http://servicenotes.literature.keysight.com/litapp/SearchSN.do?method=openExternalSNSearch&prodNum=34970A (http://servicenotes.literature.keysight.com/litapp/SearchSN.do?method=openExternalSNSearch&prodNum=34970A)

Would be nice if you could download and burn your own PROM.
Jay
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on January 05, 2016, 03:21:56 pm
Found following thread about FRAM exchange
https://www.eevblog.com/forum/testgear/agilent-34970a-errors-and-battery-replacement/ (https://www.eevblog.com/forum/testgear/agilent-34970a-errors-and-battery-replacement/)
And buch of others with usefult information for 34970A restoration
https://www.eevblog.com/forum/testgear/hp-34970a-error-741-fix/ (https://www.eevblog.com/forum/testgear/hp-34970a-error-741-fix/)
https://www.eevblog.com/forum/testgear/hpagilent-34970a-teardown/ (https://www.eevblog.com/forum/testgear/hpagilent-34970a-teardown/)
https://www.eevblog.com/forum/testgear/hp-34970a-errors-740-741-747-748/ (https://www.eevblog.com/forum/testgear/hp-34970a-errors-740-741-747-748/)
Thanks Free_Electron
So ordered two pieces  of  Cypress FM24C04B-G and hopefully I will replace also 27C4096 OTP EPROM with firmware and K102 relay.
Latest still seems to be v13.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: plesa on February 13, 2016, 09:09:03 am
Scored on 34901A (  20 Channel Multiplexer (2/4-wire) Module)  this is the most universal module for 34970A.
Started with check according to service manual and found that five relays does not work (K107-K111).
According to odometer all relays has about 500k cycles, which is minimal relay life ( electrical ).
Contact resistance on relays is according to spec in manual ( < 2Ohm).
In the service manual for 34970A is schematic for module http://cp.literature.agilent.com/litweb/pdf/34970-90012.pdf (http://cp.literature.agilent.com/litweb/pdf/34970-90012.pdf) on page 245.
Suspected IC was immediately 74HC574 ( U104  and U103)  and U104 it is dead according to IC tester. Logo on IC is Phillips, manufactured in 2003, so Agilent needs to have plenty of them, module has been manufactured in 2010 ( firmware 2.3 ) :)
uC soldered directly to board without socket. Another strange things is that the 20 channels relays are made by Panasonic TX2 and not by Omron G6SU like on previous version of module.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: douardda on June 02, 2016, 07:46:53 pm
Hi,

I just  bought a sick 34970A (with a 901 and a 907 for 200€ with shipment).

It has a sick display (at least one of the 2 flv drivers is dead, and the display itself is a quite dim). I'm waiting for the replacement parts, meanwhile I'm playing with it via the GPIB interface (I'm trying to implement a sigrok driver for it).
 
And it gives enough 700 errors (including several DMM cal data lost) each time I power it up that I can tell its 2 FRAMs are dead.

It also have a very old firmware revision (8-1-2), so since I am in the process of trying to give it a chance for a second life, I'd like to upgrade its firmware.

Has anyone dumped a recent firmware?  plesa have you found a v13 dump somewhere?

Thanks,

David

   
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on June 02, 2016, 07:50:10 pm
Keep us updates with your repair and show some pictures.

I have an older version and a newer version.
May be I can get the FW versions and read them out for you.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: douardda on June 18, 2016, 12:20:10 am
Hi

so my attempt of repairing my 34970A is not going so well so far.
I've replaced the 2 VFD drivers (putting them in PLCC sockets instead of soldering them directly).
(https://whatever.sdfa3.org/images/hp34970a/front-panel-sockets.jpg)

But the VFD is not reacting well to my rejuvenation assaults.
(https://whatever.sdfa3.org/images/hp34970a/front-panel.jpg)

With the VFD being dead, I was thinking about hacking the device by replacing the VFD display with a blue OLED display managed by an arduino or a nucleo that would take place of the 2 shift registers (in the VFD drivers) to decode what to display.

But I'm afraid I killed (don't know exactly how, I did nothing specially wrong when it started to fail) the U1 microcontroller of the display board yesterday evening since it does not give me any more signal on SEGDATA (U1 pin 5).

From there, I see 2 options:

- I find a replacement front-panel unit (they seem to be still available from Keysight for a bit more than 200€)
- I inspect the serial communication to rewrite a implement a new front-panel management controller.
  I've looked a bit at what happens there, and it should not be too difficult.

I don't event want to try to dump the ROM from U1, since being a 87C51, it's most probably "encrypted"
(XORed with a 32bit key). And on my unit, if P1.3 pin is dead, I guess I won't be able to dump the
ROM (unless it still work as an input, sice P1 is used to set the address when dumping the ROM). I may give a try. I don't know yet.

The good news is that I've changed the FRAM chips (both on the main board and the DMM)
and it does not complain any more about these.

I have an older version and a newer version.
May be I can get the FW versions and read them out for you.

That would be awesome! Yes please!


PS: I've written a blog post on this work on my modest blog (https://whatever.sdfa3.org).
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: douardda on June 21, 2016, 10:40:20 am
For the record, I've asked keysight if they would accept to make the ROM content of the display panel microcontroller available, but they are reluctant to respond favorably to my request:

Quote
I’m sorry but there is nothing I can do for you. The only Display part that is available is P/N 34970-66502 which price is $222.00. Keysight will not make available parts that are not set up for Trade Sale. Drivers, Computer Software and Programming Examples can be found under the “Drivers, Firmware and Software Tab at Keysight.com

I don't really understand why they can't make these kind of "firmware" available (besides the time required to find the correct file and
upload it on the support web site). Any Keysight guy is hanging around? ;-)

So now I guess I'm good for a reverse ingeeniring of the serial protocol between the 2 boards...

David


Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on June 21, 2016, 11:22:09 am
This usually is the problems with larger companies.

As you know, I have a 34970A around and could do a readout for you.
But I am not sure if it is the same version.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: douardda on June 21, 2016, 12:24:48 pm
This usually is the problems with larger companies.
yup, I did not expect too much, so I was not disappointed :-)

As you know, I have a 34970A around and could do a readout for you.
But I am not sure if it is the same version.

Yes, please! that would be very nice. I'm looking for both ROMs in fact: the display panel's 80c51
one (if it's not encrypted), and a reasonably recent version of main CPU one.

David

 
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: julian1 on June 25, 2016, 06:22:44 am
Could someone tell me if it's normal behavior for the 34970a to perform a full startup when power is applied to the unit (eg. by plugging in the power cord).

Or does it need the soft-start button on the front panel to be pressed.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: Jwalling on June 25, 2016, 09:29:24 am
Could someone tell me if it's normal behavior for the 34970a to perform a full startup when power is applied to the unit (eg. by plugging in the power cord).

Or does it need the soft-start button on the front panel to be pressed.

I've never seen one do that. Maybe there is a persistent setting that enables this? If you want to stop it from doing the full test, there's probably a way to reset the unit to factory defaults.

Jay
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: julian1 on June 26, 2016, 06:08:55 am

I've never seen one do that. Maybe there is a persistent setting that enables this? If you want to stop it from doing the full test, there's probably a way to reset the unit to factory defaults.

Jay

Thanks. It's wasn't doing a full-test, but plugging in the power cord, made it start up automatically. Without needing to also press the front-panel button.

I suspect the power-management u104 chip may have been incorrectly issuing a reset due to sensed a voltage drop when power was first applied.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: Jwalling on June 26, 2016, 10:05:23 am

I've never seen one do that. Maybe there is a persistent setting that enables this? If you want to stop it from doing the full test, there's probably a way to reset the unit to factory defaults.

Jay

Thanks. It's wasn't doing a full-test, but plugging in the power cord, made it start up automatically. Without needing to also press the front-panel button.

I suspect the power-management u104 chip may have been incorrectly issuing a reset due to sensed a voltage drop when power was first applied.

OK, I misunderstood what you were saying.
They will do that when the battery is replaced. Maybe yours is getting low? They will also do that when AC power is removed then restored if it was powered on.

Jay
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: julian1 on June 26, 2016, 10:54:14 am

They will also do that when AC power is removed then restored if it was powered on.


Good. thanks for clarifying.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: douardda on June 29, 2016, 09:21:26 pm
Could someone tell me if it's normal behavior for the 34970a to perform a full startup when power is applied to the unit (eg. by plugging in the power cord).

Or does it need the soft-start button on the front panel to be pressed.

Actually, there is an option to configure this behavior in the "Utility" menu (Shift + Advanced).

David

Title: Re: Agilent 34970A, Battery replacement and repair
Post by: gamalot on August 15, 2016, 12:08:56 pm
I have just upgraded my 34970A, using CR2302 (in holder) to replace the original CR14250 battery  :)
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: carl_lab on August 15, 2016, 07:03:40 pm
Which lifetime do you expect using your CR2032?

Capacity of CR2032 (230mAh) is only 1/4 of CR14250SE (~900mAh)...

http://www.varta-microbattery.com/applications/mb_data/documents/data_sheets/DS6032.pdf (http://www.varta-microbattery.com/applications/mb_data/documents/data_sheets/DS6032.pdf)
http://www.batteryjunction.com/varta-cr12aa.html (http://www.batteryjunction.com/varta-cr12aa.html)

Title: Re: Agilent 34970A, Battery replacement and repair
Post by: Jwalling on August 15, 2016, 07:22:43 pm
Which lifetime do you expect using your CR2032?

Capacity of CR2032 (230mAh) is only 1/4 of CR14250SE (~900mAh)...

http://www.varta-microbattery.com/applications/mb_data/documents/data_sheets/DS6032.pdf (http://www.varta-microbattery.com/applications/mb_data/documents/data_sheets/DS6032.pdf)
http://www.batteryjunction.com/varta-cr12aa.html (http://www.batteryjunction.com/varta-cr12aa.html)

I did some measurements a few years ago, and one could expect at least a year or so.

Then figure:
1. Ease of repair - no de-soldering required and no stressing the same solder pads over and over. Just pop a new one in. If you're careful, you can do it without losing your setups and logging memory with power applied.
2. Cost - about 30 cents (US) versus $5 or more.

Jay
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: gamalot on August 16, 2016, 04:58:47 am
Which lifetime do you expect using your CR2032?

Capacity of CR2032 (230mAh) is only 1/4 of CR14250SE (~900mAh)...

http://www.varta-microbattery.com/applications/mb_data/documents/data_sheets/DS6032.pdf (http://www.varta-microbattery.com/applications/mb_data/documents/data_sheets/DS6032.pdf)
http://www.batteryjunction.com/varta-cr12aa.html (http://www.batteryjunction.com/varta-cr12aa.html)

I did some measurements a few years ago, and one could expect at least a year or so.

Then figure:
1. Ease of repair - no de-soldering required and no stressing the same solder pads over and over. Just pop a new one in. If you're careful, you can do it without losing your setups and logging memory with power applied.
2. Cost - about 30 cents (US) versus $5 or more.

Jay

 :-+
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on April 20, 2017, 01:19:28 pm
I was lucky to get a 34972A very cheap, because it was "broken" and I expected another repair.
This is the successor of the 34970A with LXI and LAN and USB connection.

Turns out to be that the display did not turn on, because it was turned OFF
The instrument works and looks like new, even the fan had no dust on it.
This unit was built in 2012 and even the calibration certificate was online at the Keysight website.
So, in other words, NO repair needed.

Now I took the opportunity to look inside and take some pictures.
This new model looks very different in the inside, compared to the older models.
Mostly noticeable, a new battery holder at the bottom of the unit.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on April 20, 2017, 01:24:29 pm
Pictures of the inside
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on April 20, 2017, 01:29:05 pm
More pictures of the bottom
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: nfmax on April 20, 2017, 01:30:32 pm
Oh wow! I was only half serious when I suggested checking if the display was turned off, and that turned out to be the problem after all.

I gather the 34970A is basically a 34401A type DMM + scanner, while the 34972A is a similar scanner but based on a later DVM - the 34410A, I think (never had either of these in my hands to look)
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on April 20, 2017, 01:35:15 pm
Oh wow! I was only half serious when I suggested checking if the display was turned off, and that turned out to be the problem after all.

I gather the 34970A is basically a 34401A type DMM + scanner, while the 34972A is a similar scanner but based on a later DVM - the 34410A, I think (never had either of these in my hands to look)
Yes, you were right with the display, that was the first I checked.

That is an interesting suggestion you make there, that this 34972A model is based on the 34410A.
I will have a closer look.
It should be easy to find out, since the 34410A was much better in low voltage AC measurements, than the 34401A
That will be an interesting test.

Interestingly, the bootup time takes a little longer than the older models.
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on April 20, 2017, 02:51:55 pm
Unfortunately this is not confirmed...

It seems the 34972A is also using the circuitry of the 34401A, at least when it comes to low AC voltage measurements.
It is a known fact that the 34401A has problems to accurately measure low AC voltages

Here I did a test with 250 mV AC, 400 Hz
This is the lowest setting of my calibrator.

and these are the measurement results:

- 34401A:  54 uV, jumping around +/- 8 uV
- 34410A:  250.8 uV, stable
- 34411A:  253.7 uV, stable
- 34970A:  53.4 uV jumping around +/- 8 uV
- 34972A:  52.2 uV jumping around +/- 8 uV
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: nfmax on April 20, 2017, 03:27:24 pm
Well that is interesting. I wonder why they did such a major redesign on the 34972A if they weren't updating the DMM? I've always preferred the 34401A over the later 6 1/2 digit DMMs because it doesn't have a fan, but there does seem to be a real performance gain
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on April 20, 2017, 03:41:43 pm
Well that is interesting.
I wonder why they did such a major redesign on the 34972A if they weren't updating the DMM?
- Removing the GPIB interface
- Add LAN
- Add USB
- Add LXI
And probably solving lots of problems with the older design like with the battery and FRAM.

But I am surprised as well, of how much effort was put in to the re-design.


Quote
I've always preferred the 34401A over the later 6 1/2 digit DMMs because it doesn't have a fan, but there does seem to be a real performance gain
Yes, I agree.
And even with some 34410A and 34461A meters around, I like to gram the 34401A for most of my daily measurements.
May be just habit.
The fan noise of the 34410A can be reduced with a lower fan speed / different fan.

Back to the 34972A
It seems the screen update rate is also higher on the new model, compared to the old one.
 
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: nfmax on April 20, 2017, 03:46:49 pm
Quote
I've always preferred the 34401A over the later 6 1/2 digit DMMs because it doesn't have a fan, but there does seem to be a real performance gain
Yes, I agree.
And even with some 34410A and 34461A meters around, I like to gram the 34401A for most of my daily measurements.
May be just habit.
Heh! It took long enough to wean me off my 3478A onto a 34401A. Having the ranges running up to 3xxx means there is often no significant resolution gain on the 344401A - though the accuracy specification is better
Title: Re: Agilent 34970A, Battery replacement and repair
Post by: HighVoltage on April 20, 2017, 04:55:54 pm
Actually on the 34401A as well as the 34970A / 34972A you can gain one more digit and make it a 7 1/2 digit meter, if you read out the data remotely. It is just probably not very useful.