Agilent 53132A Counter Repair (FIXED and Upgraded)
HighVoltage:
The "upgrade" parts arrived from Poland and got installed today.
After turning on the instrument it gave the message "Not Calibrated", so it detected the 10 MHz OCXO alright.
The 3 GHz Pre-Scaler also seems to work.
Here are some pictures of the installed upgrades.
Next is to calibrate and check all specifications.
WN1X:
:-+
sorenkir:
Hi,
I have replaced the noisy fan of my HP53181A with the following: Sunon A40201V4 (Super Silence Series :))
http://www.digikey.com/product-search/en?mpart=HA40201V4-000U-999&vendor=259
The fan is very quiet by itself (12.8dB-A), but when installed, due to airflow & high rotation speed, it is still noticeable in a quiet environment.
Michel.
HighVoltage:
Thanks for the Digikey link to the fan.
I found a similar fan in Germany and installed it.
It is very quiet and in the lab I do not hear the noise of this instrument.
Calibration was too easy with a good 10 MHz source.
Instrument is finish now.
All done now.
Thanks to all that helped me with this one.
Vgkid:
Is that fan strong enough, 5.5 cfm is not much.