HighVoltage:

--- Quote from: texaspyro on December 27, 2017, 04:41:23 pm ---Where are you getting the VFDs?

--- End quote ---

Here is a picture after calibration with the new VFD installed.
What a difference!
It is like new from a brightness and sharpness point of view.

After searching forever, I got very lucky to get two brand new VFDs but unfortunately the seller does not have any more.

glarsson:
My Agilent 53132A with factory HS oven (option 010) does not have the extra power cable. The oven is powered through the flat cable only. The service manual mentions that the extra power cable isn't always used. Perhaps only when the oven requires it (depending on manufacturer/model number?).

HighVoltage:

--- Quote from: glarsson on December 27, 2017, 05:25:21 pm ---My Agilent 53132A with factory HS oven (option 010) does not have the extra power cable. The oven is powered through the flat cable only. The service manual mentions that the extra power cable isn't always used. Perhaps only when the oven requires it (depending on manufacturer/model number?).

--- End quote ---

Very interesting.
My two other 53131A both have the extra power cable from the power supply to the 010 option board.
Slightly different part numbers on the crystal oscillator:
10811-60160
10811-60161


glarsson:
The HS oven in my 53132A is 10811-60160. The PCB it is mounted on does not have P2 mounted (the coax connector for the blue wire) but it does have J2 mounted (for the power supply) but plugged into it is a connector that only connects the two pins closest to P2 together.

glarsson:
Image of P2 and J2.