Author Topic: Agilent 53151A Microwave Frequency Counter  (Read 582 times)
Offline Marco_MarTopic starter
« on: January 27, 2024, 08:45:38 am »
Hi Group,
need your help, please.
This unit is spot on as far as self-test and frequency measurement on both channels is concerned BUT it does not make any power measurements on channel 2.
Whatever level (according to spec of course!) of power I fed in, the unit always read -42 dBm.
Visual inspection inside the unit showed nothing remarkable.
I could not find any help with this issue surfing the net. The service manual is utterly useless.
Any hint?
Many thanks
Offline TERRA Operative
  • Country: jp
  • Voider of warranties
    • Near Far Media Youtube
Re: Agilent 53151A Microwave Frequency Counter
« Reply #1 on: January 27, 2024, 04:06:58 pm »
Did you find the CLIP?
Schematics should help a little. :)
I also added the available service notes, just in case they help...
« Last Edit: January 27, 2024, 04:09:32 pm by TERRA Operative »
« Reply #2 on: January 27, 2024, 05:55:30 pm »
Thank you but my counter is 53151.
Do you happen to have CLIP for this unit?
Bye
  • Country: jp
« Reply #3 on: January 29, 2024, 08:03:26 am »
Ah, whoops... I've been playing with the 53131A/53132A/53181A instruments recently and totally misred the nunber.... :D