Hey guys,
I recently acquired an old broken scope, an Agilent 54825A (4ch 500MHz, independent 2Gs/s sampling) , VIN # 033. It has all the extras like LS-120 drive, sound card, LAN card, GPIB, and has the acquisition board with the hybrids instead of the ceramics. Anyway, it'd be worth the repair even without all the extras, so if anyone could shed any light that'd be great.
Anyway, it came to me in great physical condition, but with what seemed to be a software bug. The waveform display area was solid orange. Still had a UI, menus worked, relays clicked when changing vertical scales, just always with a solid orange in the area where waveforms should be.
Easiest fix was software, so I followed this guide for recovery (using the external CDROM drive method):
http://pdub.net/projects/agilent-recovery-cd/ . Which had no effect. The orange seems to come from the default marker colour in software, changing the marker colour in menus will change the colour of the fill on the screen.
After reading the service guide, figured out that the windows UI is overlaid with the waveform display by the VGA mux, I started some hardware debugging. Seems to be that the display board is working fine, as the windows UI gets through, and the card can display both on the internal LCD and an external VGA monitor. The acquisition board also seems to be fine, as there is lots of high speed digital data coming across the W11 ribbon cable (acquisition->interface). However W14 (interface->display) seems to have no data, probing it out shows 2 clocks, and the rest of the lines are either held high or low.
So to me this seems like the interface board is faulty, it's receiving data, but not sending any data to the graphical buffers on the display board. However I'd like your opinion on this, as I'd like to be pretty sure of the diagnosis before dropping $500 on a replacement interface board.
If you had any other ideas or suggestions that'd be also very appreciated. Thanks for your time.