Author Topic: Agilent DSO-X 3000a (3034a) Repair issues  (Read 616 times)
Offline sahko123Topic starter
  • Country: ie
« on: May 19, 2023, 11:24:57 am »
As seems to be a very common problem my DSO-X 3034A doesnt work......
The VP23V line which ends up being the lcd bias line is at 0V which may, or may not be a consequence of the NAND corruption and a low enable line. I wanted to know if keyisght were still accepting the NAND corruption for free repair?
And if not how would i go about repairing the firmware?
The full list of symptoms is,
1. VP23V is at 0v
2. The front panel LEDs come on and flash once and then just stay on in some slightly random pattern.
3. the front panel pcb is receiving 0.7v on the 13v line which i sent over through a mosfet which isnt turned on.
4. The scope doesnt boot at all and isnt detected over usb.
Asking for a friend
Offline m k
Re: Agilent DSO-X 3000a (3034a) Repair issues
« Reply #1 on: May 23, 2023, 08:07:48 am »
One would expect that in this case a machine with corrupted memory is not random.
Random here is more like no memory.
How are other voltages?
Advance-Aneng-Appa-AVO-Beckman-Data Tech-Fluke-General Radio-H. W. Sullivan-Heathkit-HP-Kaise-Kyoritsu-Leeds & Northrup-Mastech-REO-Simpson-Sinclair-Tektronix-Tokyo Rikosha-Triplett-YFE
Offline HighVoltage
« Reply #2 on: May 23, 2023, 09:12:46 am »
Best to contact Keysight
As far as I know, they are still accepting NAND corruption repair free of charge.
But you have to send the scope first to Keysight for inspection.