I recently picked up a broken E3648A power supply that doesn't power on. Since the schematics are available online, and I have a working E3648A unit, I figured this one shouldn't be too hard to repair.
Unfortunately, it seems that the fault is inside the transformer, which is a beast. The primary side has two windings, white-black and yellow-red-orange (red is center tap), to support 100V, 115V, or 230V operation. Since I'm in the US, which is 120V, the two primaries are connected in parallel, white-black and yellow-orange. In the schematic below, both switches S100 and S101 are in the up position for 120V. When I measure resistance, there is no connection between white-black and red-orange; only yellow-red is still connected. Of course, this means that unit is completely inoperable, because even for 100V or 230V operation, the first white-black winding is connected in series to the second winding.

My next thought was to measure the secondaries, to see if I could buy an equivalent transformer. It turns out that there are nine secondary windings, of which six have center taps. The voltages aren't labeled on the schematic, but since I have a working unit, I can just measure them myself. Not including the center taps, these voltages are 27, 27, 29.4, 29.4, 38.8, 22.4, 8.8, 8.8, and 5.8. Crap, this is a completely custom transformer.
Hmm, so perhaps I can get a replacement from Agilent. The part number is E3648-60013, and although all four models of the E3646A - E3649A series are very similar, the transformers appear to be different for each unit (part numbers E364x-60013), probably to save cost. The plastic bobbin for the primary does have E3648A/E3649A printed on it, so this appears to be reused between the two units, but the actual winding might be completely different. But, E3648-60013 doesn't show up in their online parts store, so I doubt it's something they'd sell directly. There's also not much printed on the transformer, so I don't know if these are from an outside vendor.
I briefly also considered rewinding the transformer myself, but the whole thing is completely coated in varnish, and given the complexity of the transformer, I doubt that I can disassemble and repair it without breaking something. The transformer also appears completely fine; there are no signs of charring, scorching, or any sort of damage. And, the leads appear fine, up until they enter the transformer assembly. For reference, here's a photo.

The whole situation is unfortunate, because cosmetically, the unit is in great condition. Plus, when I temporarily swapped the transformer from my working unit into this one, it works fine; passing self-test and regulating the output. So, I'm stumped, because I don't know how to just repair or replace the transformer. Does anyone know how much Keysight would charge to repair the unit? Or any other suggestions?