Hi, I was recently given an agilent infiniium MSO8104A scope that is having some issues. When the scope powers up all of the front leds and fans come on and that is about it. The screen stays off and the unit constantly beeps about every 3 seconds. I think its getting stuck passing POST? I have checked the power supply and everything seems to be normal except the post labeled +/- 12.5 have +/-12.1 respectively, I think that is normal? Is this scope worth spending some time on or is it gone?
Thank you!
Fixed: After reseating RAM and CPU scope turned on but gave errors, see below. Updated the drivers, IVI shared components and IO library. One of them failed to install but upon trying to install a second time worked just fine. Restarted scope and still had the same errors. Updated to the newest version of Infiniium, also failed to install the first time, worked the second time trying to install. Newest version of Infiniium took away both errors, scope is up and working great now.