 Hello I recently received an Agilent U1620A for repair.
Channel 2 of the scoop does not work.
The green line does not appear on automatic triggering.
If you do the self -calibration function, an error code will appear channel2 offset: 0x8ff.
Channel 1 is doing well and the meter function.
I requested a service manually with schemes at Keysite, but these gave a very typical answer.
That with this cheap (7000 euros mind you) handheld, equipment does not deliver manuals with schedule.
And that I have to offer the device to them for repair.
Can anyone help me with schedules or block schedules or has some one a different idea to repair this oscilloscope.

Title: Re: Agilent U1620A Channel 2 not working
Post by: m k on June 12, 2023, 02:47:14 pm
There are two channels, so you can expect two pretty equal routes in.
If nothing else is available you can always start comparing.