Thanks for your response. I should have been more clear, as not much can be assumed about my knowledge as a newer member of this forum.
They all seem to have the same issue? That usually suggests operator error.
I have three of these loads. One of them shows the correct V/I measurements on the display, but the other two are erroneous- one shows zero voltage no matter what, the other shows like 30% off of what it should.
Assuming it is in CC?? it is hard to tell...what is the setpoint? If it is in CV, what is the setpoint?
Those pictures were all in CC mode with a setpoint of 0.1A on the load. I found out I only get the sporadic "UNR" error when the load is set to <0.5A because the load gets confused, being that it is always off by ~0.5A (more on that below). When is set it to higher currents, it seems to work indefinitely.
Do you know what the buttons do? Do you know how to use an electronic load?
Yes. I have regularly used them in the past, just not these fossils.
Is it dynamic or static?
Static.
Did you try a different mode of the load? There are CV CC CR CP?
Yes, both CV and CC do not work on the most faulty load (didn't test CR or CP); when placed in CV mode, the load pulls no current from the PSU. When in CC mode, the faulty load always pulls 0.5A more current than it is set to. All modes work, though for the wrong values, for the suspected out of cal unit. And they work as expected on the functional load.
I noticed the faulty load either displays 0.0001v or 0.0056v, never any other value. Perhaps the the ADC is fried? Or maybe it only has like 2.5 digits of accuracy instead of the shown 4 digits and is just fluctuating between the closest integers.
I've not tried remote operation as the front panel seems to work fine, and local control is how I normally would use the loads. I attached your requested photo of the rear as well as a photo of what the working unit shows on the display when in CC mode with a 1A load setting.
Thanks